Sciencemadness Discussion Board
Not logged in [Login ]
Go To Bottom

Printable Version  
Author: Subject: It lives! (HP 5890 gas chromatograph)
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 20-1-2018 at 16:23
It lives! (HP 5890 gas chromatograph)


Well gentlemen, the struggle is real, but it's up and running!

A few preliminary findings:

It was sold as 'fully tested and working'. Anybody who says that on eBay is lying to you; what they really mean is they turned it on and it didn't kick out an error message. The good news is that if they claim it works, they're on the hook for the things you find that are broken. In my case, a gas line was broken on one of the injection ports; the seller gave me a hundred bucks back to buy a replacement.

Curiously, eBay is offering an Extended Protection Plan for these things. For $30 they were promising to either fix anything that breaks over the next year or refund my purchase price. It's something to consider if the purchase makes you nervous.

The 5890 came in amazing range of power options, but mine (and most that I've seen for sale in the US) use 110 volt. My unit draws 1,400 watts at full gallop, which won't be a problem for most household circuits as long as you aren't running much else in the room at the same time. :-) It has a 'special' 20 amp plug (with one blade at a right angle to the other) but Amazon had a $10 adapter that let me plug it into a normal outlet.

This unit has flame ionization detectors (FID), which means it has FOUR separate gas inlets! The valve banks are clearly labelled, so it wasn't too hard to figure out what was what.

To power the flame inside the FID you need clean compressed air (40 psi) and hydrogen gas (15 psi.)

There's also a 'makeup gas' port that takes nitrogen (40 psi).

And then there's the carrier gas that actually flows through the column. Hydrogen or helium is recommended. Since I already had a hydrogen tank for the FID, I just used hydrogen for my carrier gas as well. (They suggest a min of 50 psi.)

So that's three gas cylinders (hydrogen, nitrogen, air) with regulators. Or two and an oilless air compressor. (I had a scuba tank and regulator handy so I used that for my air supply, using a cheap inline regulator to reduce the pressure from the scuba reg.)

The carrier gas supply (hydrogen in this case) needs a filter to remove oxygen in order to protect the column. All gas supplies need hydrocarbon filters.

Gas cylinders can be leased from your local welding/industrial gas supplier. They come in a lot of sizes; I went with the "Q" size, which seems to be a good compromise between capacity and movability:

https://www.praxairdirect.com/Industrial-Gas-and-Welding-Inf...

All the plumbing goes together with 1/8" copper tubing with compression fittings. These are common and fairly cheap. My regulators came with an outlet for some sort of hose, but the outlets themselves were standard 1/4" NPT (pipe thread), so I was able to simply screw them out and screw in an adapter that went from the pipe thread regulator port to 1/8" copper tube. You'll need plenty of teflon tape on any threaded connectors!

For regulators, I like the Victor Edge design, but anything will work as long as it has the right connector type. Gas cylinders come with about a hundred slightly different connectors to prevent the wrong thing from getting hooked up the wrong way, so you can't buy a cheap oxygen regulator and slap it on your hydrogen tank.

Hunting leaks was a bit of a pain. I ended up getting some soapy 'leak detector' fluid in a little spray bottle from the gas section of my local home store, which worked well enough (messy though.) A hydrogen leak in the oven could end very badly, but there's really just the one connection to worry about (where the column connects to the injection port.)

There are plenty of good youtube videos on installing columns. You'll need some small (1/4" and 3/16") wrenches.

Once all the gas lines are hooked up, the easy way to check for a significant leak is just to pressurize the system, turn off the gas tank, and see how long it takes for the pressure in the system to drop.

All the plumbing makes these things hard to move, perhaps on par with a large refrigerator. Give some thought to where you want it. The GC itself is about 28" wide, and needs plenty of room in the back for venting hot air from the oven. I have mine on a 48"x30" stainless commercial kitchen table, which works nicely.

Getting the unit running has been a seemingly endless game of 'find the next small part', but the availability of parts is excellent and prices are low. In particular, you'll likely need a sample syringe (perhaps 10 microliter), 11 mm rubber septums (they're a consumable), a column, a replacement injection inlet liner, a seal for the liner, ferrules to seal the column in place (another consumable), and perhaps a hanger to hold the column in the oven. (The latter can always be cobbled together with some stiff wire if you prefer.)

Manuals are readily available online. There are three (that I know of.)

The site prep manual: https://www.agilent.com/cs/library/support/documents/a15282....

The regular user's manual: http://photos.labwrench.com/equipmentManuals/128-6712.pdf

And the service manual: http://ipes.us/used/58901.pdf

There are a number of cheap old gas chromatographs out there, but for me the 5890 is a clear winner because of how common they were; it's fairly easy to find information and parts.


So what has this beautiful abomination cost?

$600 for the chromatograph (shipped.)

About $1,500 for gas cylinders, regulators, supplies, and random parts.

That puts this adventure in the ballpark of a two grand ask, and that's buying a lot of stuff used on eBay. Don't expect to get away for much less than that!

You should also plan on spending a couple months on the project, adding up to perhaps a full work day or so as you slowly figure out what you need and gather the bits and pieces.

But, hot damn! I have a working gas chromatograph!! :-)

It's a sexy, sexy piece of hardware; it'll give you the tingles in your nethers to listen to 'er hum and whir. Construction is rugged, with things generally being very repairable and accessible.

Based on a test with some butane gas from a lighter, it appears the detection threshold is probably around 0.1 micrograms as-is! Signal is a bit noisy; I'm guessing it will improve in time as contaminants from the setup bake out. But for my purposes it's just fine; even +/- 1% accuracy is plenty for me.

Here's an early test run using 1 microliter of vanilla extract from the kitchen (the first peak is alcohol, the second is the vanillin):

gc5890.jpg - 121kB gc-vanilla.jpg - 63kB

regulator.jpg - 151kB

[Edited on 21-1-2018 by Reboot]

[Edited on 21-1-2018 by Reboot]
View user's profile View All Posts By User
j_sum1
Administrator
********




Posts: 6325
Registered: 4-10-2014
Location: At home
Member Is Offline

Mood: Most of the ducks are in a row

[*] posted on 20-1-2018 at 17:16


Woohoo!
That is a lovely piece of kit. Quite an investment in both time and money even if it was a bargain. But congratulations. I am sure you will get a lot of good use out of it.




View user's profile View All Posts By User
Mesa
Hazard to Others
***




Posts: 264
Registered: 2-7-2013
Member Is Offline

Mood: No Mood

[*] posted on 20-1-2018 at 20:57


Wow nicely done. That unit looks much fancier than the one we had at TAFE(It's got an LCD display!:P)
View user's profile View All Posts By User
JJay
International Hazard
*****




Posts: 3440
Registered: 15-10-2015
Member Is Offline


[*] posted on 20-1-2018 at 21:18


Whoah awesome. I want one :)



View user's profile View All Posts By User
Texium
Administrator
********




Posts: 4583
Registered: 11-1-2014
Location: Salt Lake City
Member Is Offline

Mood: PhD candidate!

[*] posted on 20-1-2018 at 21:29


Me too, and an IR... Fortunately I have access to those, along with NMR, at school, so I don't need to buy my own machines, but it would still be really awesome to have that level of analytical prowess at home. Congratulations on getting it rebooted, Reboot!



Come check out the Official Sciencemadness Wiki
They're not really active right now, but here's my YouTube channel and my blog.
View user's profile Visit user's homepage View All Posts By User
unionised
International Hazard
*****




Posts: 5126
Registered: 1-11-2003
Location: UK
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 01:40


Very nice. You might want to keep an eye open on eBay to see if you can find a hydrogen generator.
Very high quality hydrogen without jugging gas bottles. (Also a bit safer, since it will generally shut down if there's a leak).
You can get nitrogen "generators" too- they extract it from air.
The real advantage to making your own gas is that you can get very high purity materials which improves the s/n ratio.

Commercial labs will ususaly buy 1 generator to feed several machines.
When it gets old + tired it won't produce enough gas for all of them, so they junk it.
If you only need gas for 1 GC you might get lucky and find one that will do the job.

View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3698
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline


[*] posted on 21-1-2018 at 02:06


Absolutely fabulous !

In future, when I end up in the 'WTF did I just make' situation
I know who to call :P




CAUTION : Hobby Chemist, not Professional or even Amateur
View user's profile View All Posts By User
Heavy Walter
Hazard to Others
***




Posts: 127
Registered: 17-12-2015
Location: Argentina
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 05:36


Hi Reboot

Nice job!
Looking at your vanilla extract chromatogram, seems you need to filter the FID signal.
That integrator is a poweful one, enabling to play with sensitivity in order to smooth the trace.
Congratulations for your succesful project!
View user's profile View All Posts By User
Ozone
International Hazard
*****




Posts: 1269
Registered: 28-7-2005
Location: Good Olde USA
Member Is Offline

Mood: Integrated

[*] posted on 21-1-2018 at 06:18


Just set the area reject to something above the baseline noise, say 10,000. That'll clean up your chromatogram. Then, you'll have to jigger the integration parameters to give reasonable baselines and peak cuts (splits)...if you want decent quantitative data. This is a pain in the ass using a hardware integrator. I'd recommend getting an ADAC and using a software based acquisition software package. I've had good luck using those old Dionex ACI or UI20s with unrestricted Peaknet 5.0 (IIRC) on PC.

Also, although not optimal (van Deemter), nitrogen gas can be used as a carrier. I've had good results using nitrogen at about 4.5 mL/min (1.0 is typical for He or H2; standard 30-60m column 0.25-0.5mm ID capillaries). It is both cheaper and more available than He and safer than H2 (not that I've ever had a problem with any of them).

O3




-Anyone who never made a mistake never tried anything new.
--Albert Einstein
View user's profile View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 07:54


I took apart one of the split/splitless injectors. :-)

This is the injector coming out of the GC. It's only held in place by two screws.

injector1.jpg - 146kB

Side view of the injection port on the workbench. Samples get injected into the top, the GC column connects to the bottom.

injectorontable.jpg - 191kB

The top cap removed, exposing the (red) rubber septum beneath it. These need to be replaced regularly (daily is suggested for machines that are heavily used.)

injectorcap.jpg - 205kB

The septum removed.

injectorseptum.jpg - 382kB

After carefully unscrewing the cap that the septum goes into, you'll expose the top of the liner.

injectoropen.jpg - 185kB

The liner (with it's seal around it) pulled mostly out of the injection port. These need to be replaced 'eventually' if/when they get gunked up.

injectorliner.jpg - 200kB

The bottom of the injection port. From left to right, there's a large nut that holds on the heater block, then a 1/4" to 1/8" adapter, then the column nut. To install a column, you slide it through the column nut, slide a ferrule onto the column, then tighten it together.

injectorbottom.jpg - 196kB

Inside the adapter is a gold seal with a washer. The knob on the seal points down, with the washer under it. When a column is installed, the ferrule gets jammed into the opening of that gold seal, compressing it around the column.

injectorseal.jpg - 133kB

How terrified should a person be of trying to get one of these old beasts up and running? I would say low to moderate terror. :-) There are a lot of little details, but it's a very logical machine and the manuals are freely available, so it's mostly just a matter of persistence and perhaps learning some new things (like plumbing compression fittings.)




[Edited on 21-1-2018 by Reboot]
View user's profile View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 08:32


Thinking of following me down the rabbit hole? The 5890 is HIGHLY modular and came in an amazing range of options.

There are two ways to know what you're dealing with. First, look at the panel on the left side of the machine. There are four rectangles. The top two are for the detectors. The bottom two are for the injection ports.

Looking at my machine, you can see that it has two FIDs (flame ionization detectors) because both of the upper, smaller rectangles have FID controls (three valves and an ignitor button.) They're also labelled "FID" in the upper left corner of the panels. :-)
GCsidepanel.jpg - 177kB
You'll find machines with one or two (or none!) installed. FIDs are probably the most common, but there are also photo ionization detectors (PID), Thermal Conductivity Detectors (TCD, which work by measuring how quickly the column gas carries away heat), Phosphorus/Nitrogen detectors (PND, they only detect phosphorus and nitrogen), and Electron Capture Detectors (ECD, for detecting halogens). But most units have FIDs. Machines that were hooked up to a mass spectrometer may not have ever had a traditional detector installed.

The lower two rectangles are for the injection ports. There are two common styles; simple splitless ports (most often used with packed columns) and split/splitless (more often used with modern capillary columns.) Both have a gauge, an on/off valve, and a pressure regulator knob. The split/splitless controls have an extra knob (three total) and two vents.

The split/splitless ports have the option of venting some of the sample instead of running through the column, giving you a way to dilute the sample within the machine. It's probably not that important for a home tinkerer.

For a common (30 meter, 0.25 mm ID) capillary column, set the pressure to about 12.5 psi using hydrogen or helium. Columns can cost $500-$1200 new, but you can pick up used/surplus ones on eBay for about $30 or so. :-) (I found a new old-stock 60 meter column for about $40!)

Opening the hatches on the top of the machine, you can actually see the injection ports and detectors (second photo.) Sometimes an old GC will have been stripped of parts; just because it has the controls in front doesn't prove it has the parts inside!

GCtopopen.jpg - 175kB

The tiny stainless steel lines carry gas. The woven white fabric covers the wires that control heaters in the injection ports and detectors. The temperature of each injection port, detector, and the oven itself can all be set independently.

Looking at the two FIDs in the last photo, you can see the gray wires that power the ignitors. The ignitors are just tiny heating coils, like an old school cigarette lighter. The main body of the FID is actually just stuff to create the flame and connect to the column; there aren't any electronics in it beyond the heater.)

FIDs.jpg - 161kB

Notice the silver shafts running into the sides of the FIDs. Those are important! They are the electrodes from the FID detector cards. If you don't see that shaft, the detector card is missing.

Basically the FID just burns your sample (in a hydrogen/air flame.) Burning carbon produces a tiny bit of electricity, which the electrode picks up, amplifies, and passes on to the rest of the machine.

And just for giggles, here's the mess of tanks, hoses, filters, and regulators. It's actually not as scary as it looks; just take it a step at a time.

gasses.jpg - 239kB
View user's profile View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 08:41


Thanks for all the advice, folks! Area Reject got the baseline noise under control, which is reassuring (I was wondering if I had just horribly contaminated the whole system while putting it together....)

I'll probably look for a hydrogen generator down the road; at the moment I think I've spent enough!

Computer control is definitely a long-term goal, but I must admit I take a perverse joy in using the original, slightly dinosaurish integrator. It's a bit like having a hundred year-old steam engine car that still runs well. :-)
View user's profile View All Posts By User
Texium
Administrator
********




Posts: 4583
Registered: 11-1-2014
Location: Salt Lake City
Member Is Offline

Mood: PhD candidate!

[*] posted on 21-1-2018 at 08:51


Quote: Originally posted by Reboot  
Computer control is definitely a long-term goal, but I must admit I take a perverse joy in using the original, slightly dinosaurish integrator. It's a bit like having a hundred year-old steam engine car that still runs well. :-)
I know how you feel- I think the same thing whenever I take 5 minutes to weigh something using my 60 year old Mettler balance when I have an accurate-enough-for-my-purposes digital jeweler's scale sitting right next to me. Looking forward to seeing more posts and pics on this!



Come check out the Official Sciencemadness Wiki
They're not really active right now, but here's my YouTube channel and my blog.
View user's profile Visit user's homepage View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 21-1-2018 at 10:48


Superb stuff Reboot !

It'd be great for people who have never seen/used a GC (like me !) to see how you actually run the thing, from warming up, setting gas flows etc right thru to working out what the chart means at the end.

Mega congrats on posessing a working GC.




View user's profile View All Posts By User
Dr.Bob
International Hazard
*****




Posts: 2736
Registered: 26-1-2011
Location: USA - NC
Member Is Offline

Mood: No Mood

[*] posted on 21-1-2018 at 20:00


Great job on this. Those systems are great and make analyzing solvents, lower MW chemicals, and even complex mixtures doable. If you want fun, try a shot of gasoline in it, you will get lots of peaks. Makes me laugh when I see the "Pure" brand of gas.
View user's profile View All Posts By User
brubei
Hazard to Others
***




Posts: 188
Registered: 8-3-2015
Location: France
Member Is Offline

Mood: No Mood

[*] posted on 22-1-2018 at 06:52


wonderful
View user's profile View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 22-1-2018 at 17:45



Let's stick a column in it!

The capillary columns are long, very thin glass tubes, coated on the outside with plastic and lined on the inside with a material that adsorbs chemicals. There's a wide range of options, from very non-polar coatings to very polar ones, depending on what you're working with. Relatively non-polar columns are the most common (for sorting hydrocarbons.) eBay is kind of a wonderful source for these thing; it seems like somebody is always cleaning out an old storage closet somewhere and dumping one for pennies on the dollar.


column2.jpg - 157kB

Columns come in various lengths and inside diameters. The most common size is likely 30 meters long with an ID of 0.25 mm. A common older size is 0.53 mm ID.

The columns come wrapped around a wire holder and the whole thing goes in the oven. A gas chromatograph is fundamentally just a convection oven with a way to squirt sample into one end of the column and a detector on the other end. :-) Everything else is just to support that process.


ovenoutside.jpg - 157kB

Inside the oven you might be able to make out the column (hanging from a wire column hanger). Sticking out of the top of the oven are the two injection ports (left side) and the two detectors (right side.) The big fan in the back moves air, and the heating coil is behind the metal plate between the column and the fan.

ovenopen copy.jpg - 182kB

I don't have a good photo of it, but there's a neat feature in the back: Two portholes with flaps that open and close with a stepper motor. When the GC wants to cool down the oven, it opens the flaps and vents the hot air inside. That's why you need some clear space behind the machine.

First, let's connect the injection end. Unwrap a bit of the column, slide the column nut on it, then a ferrule. The ferrules are tiny cones with a hole in the middle and are made of graphite or a graphite mix. The ferrules come in two common sizes; one fits the older 0.53 mm ID columns, while the other fits the smaller 0.25 mm ID (and similar) columns.

column-injection.jpg - 59kB

On the injection end, the goal is to end up with about 0.5 cm of column sticking out of the ferrule. You want to keep it short on the injection side because it's acting as a drain at the bottom of the injection port. If it's left long, it can act like a snorkel, sticking up above a cloud of vaporized sample and causing problems with you data. The columns are usually stored with the ends stuck into a septum (the rubber discs that seal the tops of the injection ports.) As a result, they tend to have crap stuffed in the ends and will need to be trimmed before being used. I have a $6 ceramic kitchen knife that works well for that; lightly draw the blade over the column at a right angle, then snap it off.

The column wants to slide around in the nut and ferrule until it's been swaged on. For me, the easy way to get the length right was to start with the column long, tighten the nut (swaging on the ferrule), then remove it again and trim the column back.

When you screw the column nut into place and tighten it, the ferrule gets swaged onto the column (compressed to form a permanent seal.)

column-detector.jpg - 167kB

That's kind of awkward to explain, so...go YouTube!

https://www.youtube.com/watch?v=wWnYlS0z5RQ

The detector side is left long (about 7 cm).

column-detector-loose.jpg - 105kB

On the detector side, the column needs to go inside a thin metal tube (the bottom of the FID jet). You can see it hanging down from the forward (A) detector in this pic:

detectorjet.jpg - 215kB

On the detector side, the piece that the column nut screws into simply pulls off (sealed with an o-ring I think.) On the injection port side it's a bolted on nut.

When you're tightening these column nuts up, you want to go finger tight, then perhaps a quarter turn more with wrenches.

[Edited on 23-1-2018 by Reboot]
View user's profile View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 22-1-2018 at 18:29


Speaking of that FID jet, this is what the little guy looks like:

FID-jet.jpg - 320kB

The bottom is the 'tail' hanging out of the bottom of the detector. Sample (and more hydrogen) get fed up through it. Air is supplied near the tip. (I believe the 'make-up' gas (nitrogen) is mixed into the hydrogen supply.) This is the hydrogen 'candle' that powers the detector, with the flame burning at the tip.

A wise man once said "if it ain't broke, fix it until it is!" In that spirit, let's take apart an FID! :-)

First, there are two screws holding clips onto the FID controller card's electrode (the silver shaft.) Take out the screws and slide the clips off (they wrap around the shaft on three sides; they have to come off before it will slide out.) I've already slid the control card out a little here. Notice the small spring on the end of the shaft; that's important. :-)

detectorelectrode.jpg - 195kB

Here's a shot of the right hand side of the machine, where all the electronics are housed. The two cards with gray metal boxes on them are the FID controller cards. Once the clips are removed from the electrode they can be (carefully) slid out. They fit into a computer style expansion slot.

GCelectronics.jpg - 189kB

Once you have the controller card at least partially slid out (so the spring on the end of the electrode is free) you can freely take parts off the FID detector. There's a stack of parts on top that can be removed after unscrewing that brass ring:

FIDtopparts.jpg - 196kB

Sliding the parts apart (they're loose) you get this:

FIDtoppartsloose.jpg - 138kB

Well what in the Sam Hill is all this? This...is a smokestack. It all sits on top of the FID jet (the flame.) The white (teflon) parts are insulation. The metal spool/spindle is hollow; the burning gasses from your sample rise up through the middle. The piece with the wire hanging off it just holds it all in place (the wire is for the ignited to start the flame.) That spool serves a very important purpose. Let's take a closer look at the rest of the FID first:

FIDexposed.jpg - 306kB

In the center you can see the (hexagonal) tip of the jet. You'll need a 1/4" socket or nut driver and some tweezers if you want to get it out. So, our hydrogen flame (which is burning our sample as it comes out of the column) is living right on the tip of that jet.

Now let's add back some parts:

FIDelectrodeback.jpg - 165kB

The electrode from the FID controller card is back in place; notice the spring sticking out from the inside of the brass cylinder.

Now let's grab that funny spool from the top parts:

spoolinplace.jpg - 173kB

By GENTLY setting it down inside with a bit of twisting (or just sliding the card back in place after the spool is in), you'll end up with the spring on the FID controller resting against the side of the disc on the spool:

spoolincontact.jpg - 174kB

And THAT is the key to this whole thing! The flame (rising up through this metal spool) produces a bit of electricity when carbon burns. That creates an electrical charge on the spool (which is insulated by all that teflon.) The electrode from the FID card reads the voltage off this spool and amplifies it, and that's what gives you your data. The more carbon is getting burned, the higher the voltage rises, the bigger the peak gets on the chromatograph! :-)

Fun. And very simple and robust. If one of these things was sealed up in the Great Pyramid with the Pharaoh thousands of years ago, I would expect it to still fire up and work.

View user's profile View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 23-1-2018 at 00:15


Absolutely Superb Reboot !

Truly wonderful description, detail and excellent photos !




View user's profile View All Posts By User
Yttrium2
Perpetual Question Machine
*****




Posts: 1104
Registered: 7-2-2015
Member Is Offline


[*] posted on 23-1-2018 at 00:42


I once did a gas chromatography experiment, but I don't think I fully conceptualized its uses, or benefits. Yikes
View user's profile Visit user's homepage View All Posts By User
Reboot
Hazard to Others
***




Posts: 141
Registered: 8-8-2017
Member Is Offline

Mood: No Mood

[*] posted on 26-1-2018 at 16:11


Let's talk electronics!

The 5890 motherboard reminds me how spoiled we are with USB. There are something on the order of 15 separate cable connections on it, most of which aren't used on basic models.

5890motherboard.jpg - 234kB

There are three expansion cards in this unit. To left are the two FID controller cards. To the right is the communications card.

There were apparently three communication card options for these units:

http://www.unichrom.com/hpx890/hpx890upe.shtml

The most common (and the one I have) uses INET (Instrument Network), a proprietary system using two serial cables. One cable sends data to the Integrator, the other receives commands from the Integrator.

INET.jpg - 236kB

The other two options give the GC either a traditional RS-232 serial port, or an RS-232 and an HPIB port (another proprietary connector.) You can tell if the communications card supports a normal computer serial connection by looking at the edge pointing down; if there's a ribbon cable connection there, it supports RS-232, which opens up the possibility of easy computer control and data recording. If a GC has one of the serial port capable boards installed, it should also have a serial cable connection on the back of the unit:

back2.jpg - 175kB

Looking just past the INET connectors on top there are three more connections, labelled "Remote", "Signal 1" and "Signal 2". The pins on the Remote connector can tell the unit to start a run, etc. The Signal connectors provide data from the detectors (apparently only two wires are actually needed for each since it's just a voltage reading.)

dataports.jpg - 395kB

The Service Manual has detailed pinouts for all the connectors, although it's not entirely clear which ones carry the voltage readings from the detectors. I'm sure a bored and clever person could figure it out in order to directly read the voltages coming from the detectors in order to feed them into a computer. :-)

Service Manual: http://ipes.us/used/58901.pdf

Since I've got the old school Integrator, I've been using that for capturing data. When the Integrator is connected to the 5890, they recognize each other; no further setup is needed beyond connecting the two cables and turning them on. To record the chromatograph, you hit the START button on either unit and it starts recording and printing data. To end the run, press the STOP button on either. Easy peasy.

Integrator.jpg - 205kB

The 5890 doesn't seem to support sending data from both detectors at once to the Integrator. It might be possible to read both at once from their Signal connections, which would open up interesting possibilities like splitting the sample into two columns with two different detectors on them. That way you might, for instance, be able to use an Electron Capture Detector to tell you which of your peaks was halogenated, or a Nitrogen-Phosphorus detector to tell you which peaks were amines.
View user's profile View All Posts By User
NEMO-Chemistry
International Hazard
*****




Posts: 1559
Registered: 29-5-2016
Location: UK
Member Is Offline

Mood: No Mood

[*] posted on 26-1-2018 at 16:29


Thats where that controller i mentioned comes in, it takes upto 6 outputs at once. I cant remember but I think with two detectors on that you actually use the GPIB connector to the controller, this splits the signal from each then.

GPIB was actually a standard introduced and first used by HP, it was there way of integrating different bits of there kit. I have HP logic analyzers and oscilloscopes (old ones), these are connected together via GPIB.

In the states is a electronics, second hand place (i think they on ebay), they have anything and everything HP, and at really cheap prices, if you want I will try and dig out a contact email address for you.

No point looking on there store, they got so much kit and bits of kit, only a fraction is on the site.

[Edited on 27-1-2018 by NEMO-Chemistry]
View user's profile View All Posts By User

  Go To Top