[*] posted on 3-7-2015 at 22:25
Test for Lead in Alcohol


What would be the best way to perform a test for lead in ethanol with great accuracy?

Thanks,

View user's profile View All Posts By User
Zephyr
Hazard to Others
***




Posts: 341
Registered: 30-8-2013
Location: Seattle, WA
Member Is Offline


[*] posted on 3-7-2015 at 22:52


To properly answer your question we need some more information.
How accurately?
Quantitatively?
How much are you willing to spend?
which concentrations will it be in?

I recently made some potassium rhodizonate which can be used qualitatively to test for lead...

If its as a salt in large concentrations then you could try and precipitate it and weigh it I guess.




Sciencemadness Patches for sale! U2U me if you are interested.
http://imgur.com/a/QmpHn http://www.sciencemadness.org/talk/viewthread.php?tid=62566&...
View user's profile View All Posts By User
binbin
Harmless
*




Posts: 24
Registered: 24-5-2015
Member Is Offline

Mood: No Mood

[*] posted on 3-7-2015 at 23:05


Well, as accurately as possible within $50 or so. In the end, i'd just like to see if lead is being leached from brass fittings. So a simple, yes lead, or no lead test would be great. But it would also be good to perform a titration to see exactly how much lead is in an alcohol solution.

The "great accuracy" was meant more of a, "it would be cool if.." but in the end, I'd just like to see if there is lead, or not.



[Edited on 4-7-2015 by binbin]
View user's profile View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 4-7-2015 at 04:20


http://www.docbrown.info/page13/ChemicalTests/ChemicalTestsc...

http://community.boredofstudies.org/241/chemical-monitoring-...

"generally you add Hydrochloric acid to a solution of cations. If a white precipitate forms (this is lead chloride), then lead ions are present. Judging by your question, i think you mean an additional test. To verify the lead ions, you cann add pottassium iodide and the lead ions will react with the iodide ion forming lead iodide which is a yellow precipitate "

The reaction with HCl seems to take quite a while, and at least one other reference suggests leaving it overnight to react.




View user's profile View All Posts By User
kecskesajt
Hazard to Others
***




Posts: 299
Registered: 7-12-2014
Location: Hungary
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 05:25


Well,lead chloride has a solubility of 10.8g/L.So if 5g of lead chloride formed,it cannot be detected but it is poisonus.
View user's profile View All Posts By User
Boffis
International Hazard
*****




Posts: 1867
Registered: 1-5-2011
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 05:55


Pinkhippo's suggestion that you use potassium or sodium rhodizonate is probably the best test for lead. If you evaporate down say 10ml on a white porcelain bowl you should be able to detect less than 0.1 microgram of lead. Lead gives a brownish purple precipitate and barium a reddish brown precipitate. Several other metals also give precipitates but treatment of the precipitate with very dilute (0.01M) HCl decolourises the other precipitates but turns the lead ppt to a beautiful intense violet and that of barium a crimson colour. I use this test test to not only detect Pb and Ba but to estimate the ratio of the of these metals present in minerals.

Other reagents include gallocyanine which can also be prepared easily if you have access to dimethylaniline.

@ Pinkhippo do you think it would be worth starting a thread dedicated just to the preparation and use of this type of microchemical reagent?
View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 06:11


Much simpler and more widely used than rhodizonate is simply PbS (black and very insoluble). Also PbCrO4 (yellow and very insoluble).

If you're going to concentrate the lead by evaporating off the alcohol, these tests are good enough.

On lead detection as PbS, there's a very good Utoob somewhere.

K Rhodizonate by Nurdrage:

https://www.youtube.com/watch?v=BLmycZ2nrt0


[Edited on 4-7-2015 by blogfast25]




View user's profile View All Posts By User
AJKOER
Radically Dubious
*****




Posts: 3026
Registered: 7-5-2011
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 07:23


Here is a published work out of Nigeria testing local wine/fermented berry drinks. Link: [Edit corrected] http://www.ncbi.nlm.nih.gov/pmc/articles/PMC4256578/

The wines are subject to HNO3/H2O2 pre testing to determine levels of heavy metals.

I think this work can be of use not only for the procedures employed, but also provides some data on a particular locale heavy metal levels for comparison to your sample.

Also, there is possible a statistical regression relationship between soil levels of heavy metals and your drink contents, or at least, a good testable proposition.

[Edited on 5-7-2015 by AJKOER]
View user's profile View All Posts By User
kecskesajt
Hazard to Others
***




Posts: 299
Registered: 7-12-2014
Location: Hungary
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 08:04


Quote: Originally posted by AJKOER  
Here is a published work out of Nigeria testing local wine/fermented berry drinks. Link: http://www.sciencemadness.org/talk/viewthread.php?tid=62922

The wines are subject to HNO3/H2O2 pre testing to determine levels of heavy metals.

I think this work can be of use not only for the procedures employed, but also provides some data on a particular locale heavy metal levels for comparison to your sample.

Also, there is possible a statistical regression relationship between soil levels of heavy metals and your drink contents, or at least, a good testable proposition.



[Edited on 4-7-2015 by AJKOER]

Wrong link.
View user's profile View All Posts By User
Boffis
International Hazard
*****




Posts: 1867
Registered: 1-5-2011
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 08:27


@Blogfast H2S is is very sensitive test but it is not very specific since about a quarter of the metals in the periodic table form insoluble sulphides. The chromate test is more specific but less sensitive and calcium also forms a yellow ppt and there could be a lot more calcium around than lead. To prevent interference for calcium (and Sr, Ba etc ) careful buffering is required.

Hydrogen sulphide is merely a collection reagent for groups of metals.
View user's profile View All Posts By User
binbin
Harmless
*




Posts: 24
Registered: 24-5-2015
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 08:33


I checked out Nurdrage's video. However he mentioned at the end of the lead check video that in some instances, even products he knew contained lead did not turn color upon contact. To be 100% sure, then use a professional kit or lab.

What about something like this? http://www.firstalert.com/battery/65-environmental/1033-diy-...
View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 10:29


Quote: Originally posted by Boffis  
@Blogfast H2S is is very sensitive test but it is not very specific since about a quarter of the metals in the periodic table form insoluble sulphides. The chromate test is more specific but less sensitive and calcium also forms a yellow ppt and there could be a lot more calcium around than lead. To prevent interference for calcium (and Sr, Ba etc ) careful buffering is required.

Hydrogen sulphide is merely a collection reagent for groups of metals.


Since as we're looking at brass, you'd only need to buffer against zinc, leaving on possible confusion between Cu and Pb. Which is where a chromate test could come in handy.

Rhodizonate isn't perfect either.




View user's profile View All Posts By User
WGTR
National Hazard
****




Posts: 971
Registered: 29-9-2013
Location: Online
Member Is Offline

Mood: Outline

[*] posted on 4-7-2015 at 11:31
Anodic Stripping Voltammetry for Trace Lead Detection


A more complicated solution, but if the application is important, then there are labs that run these types of tests.

Attachment: PDF-23-Experiment5.pdf (180kB)
This file has been downloaded 531 times
View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 13:35


Quote: Originally posted by WGTR  
A more complicated solution, but if the application is important, then there are labs that run these types of tests.



That method is for watery solutions. The alcohol would at least have to be diluted somewhat.

Reminds me of polarography: halcyon days.

[Edited on 4-7-2015 by blogfast25]




View user's profile View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 4-7-2015 at 13:39


Would i be a total Fool by asking if Pb always renders ions in water, or could there be undissolved Pb in the water ?



View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3696
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline


[*] posted on 4-7-2015 at 13:52


shine a laser pen through the liquid
any suspended matter will be visible
Tyndall effect https://en.wikipedia.org/wiki/Tyndall_effect
e.g. https://www.youtube.com/watch?v=vIUiK-ef8Hs

[Edited on 4-7-2015 by Sulaiman]
View user's profile View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 4-7-2015 at 14:07


Lead is said to be quite Heavy as i recall ...



View user's profile View All Posts By User
Sulaiman
International Hazard
*****




Posts: 3696
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline


[*] posted on 4-7-2015 at 14:34


give it a stir/shake before testing !
View user's profile View All Posts By User
aga
Forum Drunkard
*****




Posts: 7030
Registered: 25-3-2014
Member Is Offline


[*] posted on 4-7-2015 at 15:07


So, if it's to get water tasting better, involving shake/stir then vacuum filtering ...

"Just do the shake-n-vac to get the freshness back ..."

Anyway, how would one do a test for particulate elemental Pb in water ?

HCl again ?




View user's profile View All Posts By User
blogfast25
International Hazard
*****




Posts: 10562
Registered: 3-2-2008
Location: Neverland
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 15:32


Quote: Originally posted by aga  
Lead is said to be quite Heavy as i recall ...


It will simply sink, forming sediment. Lead in water is Pb<sup>2+</sup>




View user's profile View All Posts By User
Zephyr
Hazard to Others
***




Posts: 341
Registered: 30-8-2013
Location: Seattle, WA
Member Is Offline


[*] posted on 4-7-2015 at 16:37


Quote: Originally posted by Boffis  

@ Pinkhippo do you think it would be worth starting a thread dedicated just to the preparation and use of this type of microchemical reagent?


Yes that would be great if others want to experiment with similar tests, I know I would. It would be nice to collect a large array of similar tests so in the event of a false positive/negative it could be crosschecked to establish greater accuracy.

I prepared mine with great success using Nurdrage's method.




Sciencemadness Patches for sale! U2U me if you are interested.
http://imgur.com/a/QmpHn http://www.sciencemadness.org/talk/viewthread.php?tid=62566&...
View user's profile View All Posts By User
binbin
Harmless
*




Posts: 24
Registered: 24-5-2015
Member Is Offline

Mood: No Mood

[*] posted on 4-7-2015 at 18:04


On the distilling forums, there is a tale going around that if the brass fitting containing lead is in the lower part of the column, the lead will not be carried up by the alcohol vapours to the condenser. However if the brass/lead issue is after the head of the column, the lead will continue on to the distillate.

Any truth in this?

View user's profile View All Posts By User
macckone
Dispenser of practical lab wisdom
*****




Posts: 2168
Registered: 1-3-2013
Location: Over a mile high
Member Is Offline

Mood: Electrical

[*] posted on 4-7-2015 at 20:05


Quote: Originally posted by binbin  
On the distilling forums, there is a tale going around that if the brass fitting containing lead is in the lower part of the column, the lead will not be carried up by the alcohol vapours to the condenser. However if the brass/lead issue is after the head of the column, the lead will continue on to the distillate.

Any truth in this?


Lead is non-volatile so it won't distill but you could get small amounts of carry-over however if it is dissolved in the solution on the condenser side it will flow down. However the amount of lead in brass is usually small and the amount dissolved is even smaller. Newer brass for drinking water systems is lead free in the US.
View user's profile View All Posts By User
subsecret
Hazard to Others
***




Posts: 424
Registered: 8-6-2013
Location: NW SC, USA
Member Is Offline

Mood: Human Sadness - Julian Casablancas & the Voidz

[*] posted on 4-7-2015 at 20:40


Depending on how much you drink, you might still sustain damage even if you can't see a rhodizonate/sulfide/chromate precipitate. Soluble lead doesn't leave the body easily, and small amounts can build up over time. You'd need something better than precipitation to analyse those small amounts of lead. Lead-free brass is the way to go.



Fear is what you get when caution wasn't enough.
View user's profile View All Posts By User
binbin
Harmless
*




Posts: 24
Registered: 24-5-2015
Member Is Offline

Mood: No Mood

[*] posted on 5-7-2015 at 12:15


Quote: Originally posted by subsecret  
Depending on how much you drink, you might still sustain damage even if you can't see a rhodizonate/sulfide/chromate precipitate. Soluble lead doesn't leave the body easily, and small amounts can build up over time. You'd need something better than precipitation to analyse those small amounts of lead. Lead-free brass is the way to go.


Or no brass at all. Raw copper can be found still, it'll be more expensive and harder to find for some parts, but can be found. To be considered lead free, brass only needs to contain less than 0.25% lead, which is still not 0%.

View user's profile View All Posts By User