fusso
International Hazard
Posts: 1922
Registered: 23-6-2017
Location: 4 ∥ universes ahead of you
Member Is Offline
|
|
How accurate are solution density measurements?
I used a measuring cylinder to measure my ammonia solution's density. I calculated the density to be 0.912gcm-3. According to https://wissen.science-and-fun.de/chemistry/chemistry/densit... it should be 23.39%. But how accurate is the result? What would be the range of
the actual density?
[Edited on 190620 by fusso]
|
|
mayko
International Hazard
Posts: 1218
Registered: 17-1-2013
Location: Carrboro, NC
Member Is Offline
Mood: anomalous (Euclid class)
|
|
Way too little information to even begin: was this a single measurement, or a composite of multiple measurements? What mass(es) and what volume(s)
were measured? What size graduated cylinder did you use? (Most have a small inaccuracy near the very bottom, which means that they have a large
percent error for relatively small volumes, but a smaller percent error for large volumes.)
The phrase you probably want to look up is "propagation of error". There's also this thread; fitting a regression to multiple (mass,volume) data will give you confidence bands on the density estimated (though incorporating
instrumental error will take a little more work).
And/or, independently measure concentration with gravimetry/titration and compare results.
al-khemie is not a terrorist organization
"Chemicals, chemicals... I need chemicals!" - George Hayduke
"Wubbalubba dub-dub!" - Rick Sanchez
|
|
fusso
International Hazard
Posts: 1922
Registered: 23-6-2017
Location: 4 ∥ universes ahead of you
Member Is Offline
|
|
Single measurement, using a 25ml cylinder, 19.6g for 21.5ml.
|
|
UC235
National Hazard
Posts: 565
Registered: 28-12-2014
Member Is Offline
Mood: No Mood
|
|
Graduated cylinders are really not very accurate. Density is quite temperature-dependent. Do you know what grade of graduated cylinder you have or is
it chinese stuff? Some brands have the tolerance for the markings printed right on them.
A volumetric flask would be much more accurate. The temperature should at least be very close to 20C. A scale that has actually been calibrated, or at
least checked against a set of calibration weights matters.
|
|
SWIM
National Hazard
Posts: 970
Registered: 3-9-2017
Member Is Offline
|
|
A good set of hydrometers makes this sort of thing much easier.
I suspect even a fairly broad range hydrometer would be a more accurate than using a graduated cylinder and a scale.
|
|
Sulaiman
International Hazard
Posts: 3695
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
various notes and opinions:
. the accuracy of measuring cylinders can be Class-A, Class-B or other (usp, unclassified ...)
- you did not specify the accuracy class of your measuring cylinder
. the density of common liquids varies significantly with temperature
- you did not specify the temperature
. thermometers also have accuracy classes
- you did not specify the expected error range for temperature
. weighing scales also have accuracy limits
- you did not specify the error range for weight
You specified the density as 0.912 with no estimate of error
by default the error is +/- one last digit
so you are implying that your measurement is accurate to 0.1%
- I can't read a 25ml measuring cylinder to 0.025ml
- Class-A 25ml measuring cylinders have up to +/- 0.25ml error
- How accurate is your thermometer ?
- how accurate are your scales ?
and if you are going to specify to 0.1% or better,
then your scales need to be calibrated at your location
- due to geographical variations in gravity,
and you need an estimate of bouyancy due to your local air density !.
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
Heptylene
Hazard to Others
Posts: 319
Registered: 22-10-2016
Member Is Offline
Mood: No Mood
|
|
The density \(\rho\) of a sample is calculated from its mass \(m \pm \delta m\) and volume \(V \pm \delta V\).
$$\rho = \frac{m}{V}.$$
The uncertainty \(\delta \rho\) on the density is calculated using
$$\delta \rho = \rho \Bigg(\Big(\frac{\delta V}{V}\Big)^{2} + \Big(\frac{\delta m}{m}\Big)^{2}\Bigg)^{1/2}.$$
So you need to figure out the uncertainty on the mass and volume. For the cylinder, the uncertainty should be written on it.
|
|
DrP
National Hazard
Posts: 625
Registered: 28-9-2005
Member Is Offline
Mood: exothermic
|
|
You can get specially calibrated cups that fill to an exact volume. The lids have a hole in them so any excess can be squeezed out of the top. You
weight the contents on a decent balance (3dp?) that is accurate enough for your purposes and divide by the volume.
I guess it depends how accurate you need your measurement to be.
\"It\'s a man\'s obligation to stick his boneration in a women\'s separation; this sort of penetration will increase the population of the younger
generation\" - Eric Cartman
|
|
SWIM
National Hazard
Posts: 970
Registered: 3-9-2017
Member Is Offline
|
|
Sounds like a pycnometer.
They've got ones that include a thermometer in the design too.
|
|
Sulaiman
International Hazard
Posts: 3695
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
Forgot ... If there is enough volume of liquid then I'd use a hydrometer for quick results
with surprisingly good precision (typically much better than 1%) even using cheap hydrometers and thermometers.
To answer more directly, if I did the measurement I would expect;
+/- 0,25 ml error using a 25ml Class-A measuring cylinder
+/- 0.1 ml reading error (if I'm really careful)
+ about 0.05 ml for each drop stuck to the glass
let's say about 0.32ml error in 21.5ml = 0.15% error, approximately
(Temperature correction is required, but I'd assume that the errors in temperature correction would be less than overall errors)
after calibration I trust my cheap 300g x 0.01g scales to +/- 0.02g
so the weighing error would be about 0.1%
compensate for bouyancy (about 0.1%) with negligible error
So, I'd hope to achieve about 0.2% error in my density measurement.
Unfortunately many liquids change density slowly vs. concentration change,
so a 0.2% density error could lead to a 1% or more error in concentration calculations.
.
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
fusso
International Hazard
Posts: 1922
Registered: 23-6-2017
Location: 4 ∥ universes ahead of you
Member Is Offline
|
|
Hm, I made another measurement and get a density of 0.906(forgot the mass&vol). The cylinder is 25±0.5ml, has markings every 0.5ml but unknown
class. The cheap scale is 500x0.01g.
|
|
brubei
Hazard to Others
Posts: 188
Registered: 8-3-2015
Location: France
Member Is Offline
Mood: No Mood
|
|
all you need is in this book :
Data analysis for chemistry: an introductory guide for students and laboratory scientists
1/ Measure your density with several essay. Let say 5, so N = 5
Evaluate the mean of your density
Check the standard error
Next you can evaluate the confidence limit associate with 95% certitude with student law
where A is a value associated with a normal distribution for 5 essay and 95% confidence in table here : https://en.wikipedia.org/wiki/Student%27s_t-distribution#Tab...
A = 2.015
By this way you don't have to bother with lecture incertitude associated with your instrument (expect systematic error) because all of them are
evaluated with statistic
[Edited on 21-6-2019 by brubei]
I'm French so excuse my language
|
|
Sulaiman
International Hazard
Posts: 3695
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
Two measurements made,
1) Density 0.912, concentration 23.39%
2) Density 0.906, concentration 25.33%
Density change 0.66%, relative concentration change 8.2%
- notice the relative change in concentration is much greater than the change in density
- your two measurements differ by about 0.66%
so even if your measuring cylinder and scales and thermometer and calculations are perfect,
your own skill is limiting the accuracy of your results.
In my opinion, measuring cylinders are good enough for measuring liquid volumes for quick qualitative experiments,
I like to use either hydrometers or a volumetric flask on scales for density measurement,
and a burette for delivery volumes (e.g. deliver a measured volume into a weighing vessel)
- if I get quantitative..
A pycnometer ( I know that at least one member has one) seems an excellent choice for density measurements if you need that level of accuracy,
(a pycnometer looks like a simple diy project)
P.S. as the ammonia concentration changes much more significantly than density,
using the same equipment, a titration would be more accurate.
(IF you have an acid of accurately known concentration to titrate against, which you could make from water and weighed dry powdered acids such as
ascorbic, boric, citric, oxalic, sulphamic, sodium bisulphate .....)
[Edited on 21-6-2019 by Sulaiman]
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
Heptylene
Hazard to Others
Posts: 319
Registered: 22-10-2016
Member Is Offline
Mood: No Mood
|
|
@brubei: Your approach only takes into account the random error right?
As you said you would also need to estimate the systematic error which is due to the calibration of your instruments. Surely in an amateur setting
without a calibrated scale, the systematic error is important to consider.
|
|
BromicAcid
International Hazard
Posts: 3246
Registered: 13-7-2003
Location: Wisconsin
Member Is Offline
Mood: Rock n' Roll
|
|
Aside from class of measuring cylinder, cylinders are usually also labeled TC (to contain) and TD (to dispense) so a TD cylinder is going to have more
volume than indicated on the graduations. Just do a quick check with a known substance (water) and see where you fall.
|
|
mayko
International Hazard
Posts: 1218
Registered: 17-1-2013
Location: Carrboro, NC
Member Is Offline
Mood: anomalous (Euclid class)
|
|
Some more info on doing these calcs in Excel:
Tellinghuisen, J. (2018). Least-Squares Analysis of Data with Uncertainty in y and x: Algorithms in Excel and KaleidaGraph. Journal of Chemical
Education, 95(6), 970–977. research-article. https://doi.org/10.1021/acs.jchemed.8b00069
Attachment: Least-Squares Analysis of Data with Uncertainty in y and x.pdf (2.7MB) This file has been downloaded 594 times
al-khemie is not a terrorist organization
"Chemicals, chemicals... I need chemicals!" - George Hayduke
"Wubbalubba dub-dub!" - Rick Sanchez
|
|
CharlieA
National Hazard
Posts: 646
Registered: 11-8-2015
Location: Missouri, USA
Member Is Offline
Mood: No Mood
|
|
I have used a 10 ml Class A volumetric flask and a balance reading to 1 mg, with decent (for me, anyway) results.
|
|