Jacob
Harmless
Posts: 23
Registered: 7-4-2019
Member Is Offline
|
|
Keeping the vacuum pump clean
Well, this is rarely if ever discussed.
How do you keep solvents and volatiles away from your vacuum pump?
The oil in my rotary vane pump turns into milk every couple of vacuum distillations.
I have a glass cold trap. Maybe filling it with mineral oil and putting it in ice will work.
|
|
12thealchemist
Hazard to Others
Posts: 181
Registered: 1-1-2014
Location: The Isle of Albion
Member Is Offline
Mood: Rare and Earthy
|
|
I have no experience of using a vacuum pump outside of university (where I use them daily - Schlenk line work), so I can't help too much with the
hobby side of things.
That being said, a trap is key to protecting the pump and prolonging its life. I don't know how cold one needs to go; I've always used liquid
nitrogen, but I hear dry ice/acetone also works. If you can't access dry ice (I can't outside uni), ice and salt will reach -18°C at best, and ice
and calcium chloride hexahydrate will reach -40°C at best. I imagine this latter is good enough to catch the vast majority of your volatiles. I don't
know how good the other options are.
|
|
Jacob
Harmless
Posts: 23
Registered: 7-4-2019
Member Is Offline
|
|
Quote: Originally posted by 12thealchemist |
I have no experience of using a vacuum pump outside of university (where I use them daily - Schlenk line work), so I can't help too much with the
hobby side of things.
That being said, a trap is key to protecting the pump and prolonging its life. I don't know how cold one needs to go; I've always used liquid
nitrogen, but I hear dry ice/acetone also works. If you can't access dry ice (I can't outside uni), ice and salt will reach -18°C at best, and ice
and calcium chloride hexahydrate will reach -40°C at best. I imagine this latter is good enough to catch the vast majority of your volatiles. I don't
know how good the other options are. |
My bad. What I have is a gas washing bottle not a cold trap. Not that it makes a difference since ice and salt is the best cooling available right
now. I'll try activated carbon or mineral oil.
|
|
unionised
International Hazard
Posts: 5129
Registered: 1-11-2003
Location: UK
Member Is Offline
Mood: No Mood
|
|
Do you know what a gas ballast valve is?
|
|
wg48temp9
National Hazard
Posts: 786
Registered: 30-12-2018
Location: not so United Kingdom
Member Is Offline
|
|
In case the above comment was lost on you, below is some info about using the air ballast on a vacuum pump.
"It is possible to throughput some of these contaminants from your system and through the pump and that are where the gas ballast valve comes into
play. If you do not require high vacuum less than say 300 micron for instance, then you can open up the gas ballast valve during the evacuation
procedure and get some of these bad molecules out of the pump through the exhaust port of the pump. Remember also if they are coming out the exhaust
they potentially are entering your workplace. This may pose a hazard either health based or fire based. If you are looking for better vacuum
pressures, the valve is adjustable. This feature is not the total answer; it helps, but should not be considered the solution in problem applications
with lots of contaminates.
If and once the pump is contaminated you can again use the gas ballast valve to assist in purging some of these contaminants from the vacuum pump oil
in which it now resides. Close the pump inlet port off to full vacuum and allow pump to run/actuate with gas ballast valve in open position. This
again will help purge the pump. Take a long lunch or let the pump actuate overnight with the valve open. Care again should be considered to the
exhaust stream. If the pump is "smoking" or exiting oil mist in good quantity, you may come in the next morning to a room full of oil mist. Piping to
a hood or using a coalescing filter like the HyVac Capture filters can help eliminate this problem area."
I am wg48 but not on my usual pc hence the temp handle.
Thank goodness for Fleming and the fungi.
Old codger' lives matters, wear a mask and help save them.
Be aware of demagoguery, keep your frontal lobes fully engaged.
I don't know who invented mRNA vaccines but they should get a fancy medal and I hope they made a shed load of money from it.
|
|
Bezaleel
Hazard to Others
Posts: 444
Registered: 28-2-2009
Member Is Offline
Mood: transitional
|
|
wg48temp9, when purging the pump oil, do you recommend to close the vacuum inlet and use the gas ballast in order to avoid excessive oil mist?
Usually, immediately before I switch of my pump, I let it pump air through the inlet for half a minute or so in order to clean the inlet tubes. This
also cleans the pump's vacuum chambers.
My personal experience is that using the gas ballast is of essential value to preserve your pump. Even pumping on just water will corrode the interior
of your pump over time. Let alone if you need to pump on filtrates containing HCl and the like.
The idea behind using the gas ballast is that you allow the second vacuum chamber of your pump (the one with the really low pressure) to receive a
little air. This prevents contaminants from condensing and mixing with the pump oil, because they will remain mixed (dissolved) in the air which came
in through the gas ballast valve.
And I can confirm that the HyVac Capture filters are a must also. Since I use them my work space retains a clear sky.
|
|
DavidJR
National Hazard
Posts: 908
Registered: 1-1-2018
Location: Scotland
Member Is Offline
Mood: Tired
|
|
I use a chemical resistant diaphragm vacuum pump precisely to avoid this. You don't get as good a vacuum as a rotary vane pump, but for my purposes it
has been fine, even for Schlenk line work.
|
|
wg48temp9
National Hazard
Posts: 786
Registered: 30-12-2018
Location: not so United Kingdom
Member Is Offline
|
|
Quote: Originally posted by Bezaleel | wg48temp9, when purging the pump oil, do you recommend to close the vacuum inlet and use the gas ballast in order to avoid excessive oil mist?
|
Yes that is the recommended procedure as described in my previous post. Ideally it is done for a few hours with the pump hot.
What I used to do after evacuating a large chamber (100l) I would connected to the outlet of the high vacuum pump to a fridge compressor operating
as a vacuum pump. The high vacuum pump would have its inlet isolated. The high vacuum pump was the type that have the exhaust port connect to the
space above the oil chamber. So the fridge compressor could pull a vacuum on the high vacuum pump oil and remove any water vapour or volatile
contaminate very quickly compared to running the high vacuum pump. It was possible to observe the pump oil bubbling thru the oil level sight glass
and use the observation to determine when the bubbling had stopped and the process was complete.
I am wg48 but not on my usual pc hence the temp handle.
Thank goodness for Fleming and the fungi.
Old codger' lives matters, wear a mask and help save them.
Be aware of demagoguery, keep your frontal lobes fully engaged.
I don't know who invented mRNA vaccines but they should get a fancy medal and I hope they made a shed load of money from it.
|
|
Bezaleel
Hazard to Others
Posts: 444
Registered: 28-2-2009
Member Is Offline
Mood: transitional
|
|
That's a clever idea, wg48temp9. Thanks for posting it here!
My rotary vane pump has exactly that configuration (I think most RV pumps do) - the exhaust chamber is right above the oil level, only separated by
two plates, in order to avoid oil droplets from getting into the exhaust chamber directly. And to allow any oil collecting on the surfaces of the
exhaust chamber to flow back into the oil bath.
|
|