Vacuum pump problem
I have a rotary vacuum pump. I set up for a standard distillation of dcm and triethylamine. I put the distallation flask on a oil bath and set the hot
plate to 100c.
Now my problem is the vacuum pump is freezing the flask :s.. I have no idea why. Im trying to boil the solvent + triethylamine off but the vacuum pump
keeps freezing the flask.. Any one know why?
|
|
Readytorock
Harmless
Posts: 4
Registered: 14-3-2012
Member Is Offline
Mood: No Mood
|
|
bump!.. any one?
|
|
chemx01
Hazard to Self
Posts: 54
Registered: 13-6-2010
Member Is Offline
Mood: No Mood
|
|
I don't know if I understand you corectly but as I understand your boiling DCM and TEA under vacuum on oil bath.
Well the boiling point of DCM at this pressure could be as low as -30°C so it's probably condensing moisture on the flask and freezing it therefore
the "frozen flask".
If I didn't understand you corectly please elaborate and one advice don't rush it people usually don't take that very kindly.
|
|
garage chemist
chemical wizard
Posts: 1803
Registered: 16-8-2004
Location: Germany
Member Is Offline
Mood: No Mood
|
|
You seem to be pulling the solvent vapors straight into your vacuum pump. This will contaminate the oil and quickly destroy the pump. Immediately
disassemble the pump completely, clean it and fill it with fresh oil.
DCM and triethylamine are best removed by distillation at atmospheric pressure. If you insist on using vacuum, you will either have to put a cold trap
cooled with dry ice or liquid nitrogen between your distillation setup and the pump and have the solvent condense there, or buy a chemically resistant
diaphragm vacuum pump that lets you condense the solvent vapors at the pump outlet at atmospheric pressure and with a water-cooled condenser.
I have such a chemically resistant pump, the MZ2C from Vacuubrand, and I can only recommend that you don't use your rotary vacuum pump in the chemical
laboratory.
[Edited on 15-3-2012 by garage chemist]
|
|
Magpie
lab constructor
Posts: 5939
Registered: 1-11-2003
Location: USA
Member Is Offline
Mood: Chemistry: the subtle science.
|
|
What's the point in a vacuum distillation and a 100C oil bath to remove a low boiling solvent like DCM? It seems like gross overkill.
The single most important condition for a successful synthesis is good mixing - Nicodem
|
|
zoombafu
Hazard to Others
Posts: 255
Registered: 21-11-2011
Location: U.S.
Member Is Offline
Mood: sciencey
|
|
You don't need to be doing a vacuum distillation (unless I'm missing something) typically you only use vacuum distillation for high boiling liquids
(such as essential oils)
|
|
Readytorock
Harmless
Posts: 4
Registered: 14-3-2012
Member Is Offline
Mood: No Mood
|
|
Quote: Originally posted by chemx01 | I don't know if I understand you corectly but as I understand your boiling DCM and TEA under vacuum on oil bath.
Well the boiling point of DCM at this pressure could be as low as -30°C so it's probably condensing moisture on the flask and freezing it therefore
the "frozen flask".
If I didn't understand you corectly please elaborate and one advice don't rush it people usually don't take that very kindly. |
Thanks
|
|
Readytorock
Harmless
Posts: 4
Registered: 14-3-2012
Member Is Offline
Mood: No Mood
|
|
Quote: Originally posted by garage chemist | You seem to be pulling the solvent vapors straight into your vacuum pump. This will contaminate the oil and quickly destroy the pump. Immediately
disassemble the pump completely, clean it and fill it with fresh oil.
DCM and triethylamine are best removed by distillation at atmospheric pressure. If you insist on using vacuum, you will either have to put a cold trap
cooled with dry ice or liquid nitrogen between your distillation setup and the pump and have the solvent condense there, or buy a chemically resistant
diaphragm vacuum pump that lets you condense the solvent vapors at the pump outlet at atmospheric pressure and with a water-cooled condenser.
I have such a chemically resistant pump, the MZ2C from Vacuubrand, and I can only recommend that you don't use your rotary vacuum pump in the chemical
laboratory.
[Edited on 15-3-2012 by garage chemist] |
Thanks
|
|