I just bought the Bel-Art 329470000. I have it hooked up to a standard US faucet and I am getting next to no vacuum. I hooked the gauge directly to
the aspirator using a 60cm vacuum hose. The water output is placed directly into the drain using a 30cm PP hose. I have yet to measure a pressure of
less than 680 mm Hg. At this point I'm assuming that the aspirator is faulty but I find that difficult to believe. Is it more likely that there is a
fault in my system?
Here is the info sheet:
http://www.belart.com/shop/instruction-sheets/932947003.pdf
|
|
WGTR
National Hazard
Posts: 971
Registered: 29-9-2013
Location: Online
Member Is Offline
Mood: Outline
|
|
What is the water output in gallons per minute (measure it)?
Second, when the vacuum is being measured, are there a bunch of air bubbles going out with the water stream? If so, there is
a leak somewhere.
[Edited on 1-20-2014 by WGTR]
|
|
nolken
Harmless
Posts: 4
Registered: 20-1-2014
Member Is Offline
Mood: No Mood
|
|
2.89 GPM output without aspirator attached and 2.64 GPM output with the aspirator attached.
|
|
BromicAcid
International Hazard
Posts: 3247
Registered: 13-7-2003
Location: Wisconsin
Member Is Offline
Mood: Rock n' Roll
|
|
So that's plenty of water according to the spec sheet. Can you have too much water? Never been in that boat but I don't think so. Next question,
you're using cold water, right? Also restating WGTR's question, no air bubbles, right? If you plug both openings you don't get leaking where you're
connected to the faucet do you?
|
|
nolken
Harmless
Posts: 4
Registered: 20-1-2014
Member Is Offline
Mood: No Mood
|
|
The water exits with such a high velocity it was hard to tell so I placed the water output it in a class jar temporarily and I did not see any air
bubbles. To confirm there are no leaks I placed my finger on the vacuum port and felt no vacuum. It does make a loud hissing sound leading me to
believe the aspirator is taking in air but it may just be the water flowing through.
|
|
nolken
Harmless
Posts: 4
Registered: 20-1-2014
Member Is Offline
Mood: No Mood
|
|
Correct, I am using cold water. There are air bubbles when the vacuum port is left open so it is for sure bringing air in. upon connecting the gauge
the air bubbles stop even though the gauge is reading such a poor vacuum. I tried adjust the water flow but the change in vacuum pressure was very
little. There is no leaking when I plug the water output. The vacuum port does have a check valve that seems to be functioning properly.
[Edited on 20-1-2014 by nolken]
|
|
Dr.Bob
International Hazard
Posts: 2734
Registered: 26-1-2011
Location: USA - NC
Member Is Offline
Mood: No Mood
|
|
Does the check valve rattle when you shake the aspirator? If it does not move freely, then it will often stick. Some of them unscrew so you can
clean the ball inside.
|
|
WGTR
National Hazard
Posts: 971
Registered: 29-9-2013
Location: Online
Member Is Offline
Mood: Outline
|
|
You did leave the little plastic tube on the output, right? I think that's necessary. Also, I think the water has to occupy the whole
tube on its way out, not just part of it. Other than that, I'm out of ideas.
[Edited on 1-21-2014 by WGTR]
|
|