Vacuum regulation
Hi. My first post here after many years lurking.
I have a rotary vane pump I'm wanting to use for distillations such as high boiling oils or higher boiling solvent removal. What is the best way to
regulate the vacuum? Is it possible to get any kind of reasonable regulation by simply having a bleed valve allowing it to draw in some air? Do I
absolutely need a regulator of some kind? Are refrigeration re-gassing regulators any use? If there are any other options, could you recommend one?
All this with the understanding that my budget is seriously limited!!
The only time I've run a vacuum distillation in the past was using the inlet of an air compressor, but it didn't reach a strong vacuum and fluctuated
a lot.
Many thanks!
|
|
Sulaiman
International Hazard
Posts: 3692
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
If your reactants and products are not too sensitive
An air leak in the form of a bubbler in the boiling pot allows some range of pressure adjustability,
and prevents bumping.
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
Squibble
Harmless
Posts: 6
Registered: 9-4-2009
Member Is Offline
Mood: No Mood
|
|
OK. Now you mention it, I've heard of that before. I was thinking of a bleed in the tube to the pump, but this is better. Thanks!
When you say sensitive, you mean to the atmosphere?
|
|
Sulaiman
International Hazard
Posts: 3692
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
Yes, sensitive to components of the atmosphere.
Oxygen, water vapour, carbon dioxide etc.
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
walruslover69
Hazard to Others
Posts: 231
Registered: 21-12-2017
Member Is Offline
Mood: No Mood
|
|
My old lab used to have a comically oversized rotary vane pump connected to our rotary evaporator. When we tried evaporating anything with a boiling
point below 70C it would flash boil and bump out of control.
The solution we came up with was to connect the vacuum pump to the rotary evaporator with flexible vinyl tubing. We then placed a C clamp on the
tubing to restrict it and reduce the vacuum flow rate. sometimes up to 3 clamps were needed when evaporating solvents like DCM, but it worked
surprisingly well.
|
|
Mateo_swe
National Hazard
Posts: 541
Registered: 24-8-2019
Location: Within EU
Member Is Offline
|
|
There are vacuum regulators on some rotovaps.
Maybe you can get such a vacuum regulator, a used one.
I think they usually are a separate box with a display and sone buttons.
|
|
Rainwater
National Hazard
Posts: 919
Registered: 22-12-2021
Member Is Offline
Mood: indisposition to activity
|
|
There are two working options in this post
https://www.sciencemadness.org/whisper/viewthread.php?tid=15...
Easiest is the $70 controller.
Cheapest is a glove, jam jar and some tubing.
"You can't do that" - challenge accepted
|
|
Sulaiman
International Hazard
Posts: 3692
Registered: 8-2-2015
Location: 3rd rock from the sun
Member Is Offline
|
|
Quote: Originally posted by walruslover6 | We then placed a C clamp on the tubing to restrict it and reduce the vacuum flow rate. sometimes up to 3 clamps were needed when evaporating solvents
like DCM, but it worked surprisingly well. | just curious:
Why would you want to use vacuum with DCM, bp<40oC,
and how did you condense it?
EDIT: Thanks for the explanation you gave below. Unexpected, but logical.
........................
I like the idea of the jar, funnel and membrane regulator,
but I'd connect the regulator independently to the jar to reduce pulsing of the supplied pressure/vacuum.
[Edited on 12-6-2024 by Sulaiman]
CAUTION : Hobby Chemist, not Professional or even Amateur
|
|
Squibble
Harmless
Posts: 6
Registered: 9-4-2009
Member Is Offline
Mood: No Mood
|
|
Thanks for the ideas everyone!
|
|
walruslover69
Hazard to Others
Posts: 231
Registered: 21-12-2017
Member Is Offline
Mood: No Mood
|
|
@Sulaiman I wish I had an old picture to show, the whole system was insane. For some background This was a materials lab with a lot of money to throw
around but no experienced benchtop chemists.
The roto water bath was really over sized with an undersized heating element, so it was drastically quicker to just start the vacuum from room temp
and distill over that way. For the collection, we had a large cryogenic chiller that got down to -80C.
it was a situation where we just had tons of oversized equipment cobbled together.
Writing this now makes me look back on how surreal of a setup it was....
|
|