Problems getting a decent vacuum from aspirator station
Hey there,
I'd had an aspirator hooked up to my sink in the past for light vacuum applications, like filtering. Recently, I realized that I'd like to waste less
water, and that I'd like to be able to attempt some easy vacuum distillations.
Attached is a picture of what I've built for this.
The pump, in a plastic container with ice, sends water through the aspirator. When I connected my manometer to this (with a distillation rig on the
other end), I can only get a vacuum of -300mmHg, and I believe I should be able to get more, with the 1/2 HP pump I have.
Does anyone have any suggestions for this?
Instead of using multiple, jointed, silicone taped pipes to get to the aspirator, I should use a short hose?
Thanks for any input.
it's all about chemistry.
|
|
DJF90
International Hazard
Posts: 2266
Registered: 15-12-2007
Location: At the bench
Member Is Offline
Mood: No Mood
|
|
Your problem is that you are generating back pressure on the aspirator! your water outlet tube needs to be about 6 inches long and the end of it needs
to be ABOVE the surface of the water, in order for it to work effectively. Or at least thats my opinion on the matter.
|
|
entropy51
Gone, but not forgotten
Posts: 1612
Registered: 30-5-2009
Member Is Offline
Mood: Fissile
|
|
Not all pumps work well for driving a particular aspirator. The pressure and flow delivered by the pump may need to be matched to the aspirator.
From your drawing, it appears that you may be drawing air bubbles into the pump, which is not good. DJF is right that the aspirator outlet should be
above the water level.
There are several threads on this subject. Search using aspirator in the subject line and you will find them.
|
|
Contrabasso
Hazard to Others
Posts: 277
Registered: 2-4-2008
Member Is Offline
Mood: No Mood
|
|
What is the pump developing in terms of pressure and flow rate. You will probably need 1 - 2 bar and not a huge flow rate. You will also need to raise
the aspirator so that the outlet has a tail pipe in the order of 6 - 12 inches and this is clear of the water reservoir.
|
|
basstabone
Harmless
Posts: 24
Registered: 17-4-2008
Member Is Offline
Mood: No Mood
|
|
Another suggestion is to put a plate or a board between your pump and the aspirator which will hopefully increase it a bit. This will stop the air
bubbles from coming in
|
|
stateofhack
Hazard to Others
Posts: 123
Registered: 6-5-2008
Location: Warm Coast
Member Is Offline
Mood: annoyed
|
|
Use a baffle! No air bubbles it should do the job!
|
|
Vogelzang
Banned
Posts: 662
Registered: 26-4-2008
Member Is Offline
Mood: No Mood
|
|
Use ice cold water.
|
|
grndpndr
National Hazard
Posts: 508
Registered: 9-7-2006
Member Is Offline
Mood: No Mood
|
|
What is the effect of 115degree F tap water on an aspirators vacuum?
|
|
watson.fawkes
International Hazard
Posts: 2793
Registered: 16-8-2008
Member Is Offline
Mood: No Mood
|
|
Higher ultimate pressure. More
back-streaming of water vapor into the vacuum line, if that matters to your application.
|
|
gsd
National Hazard
Posts: 847
Registered: 18-8-2005
Member Is Offline
Mood: No Mood
|
|
@grndpndr
You either do not read the replies to your questions or do not understand the answers or both.
Reply to this very question which you raised in the "Need help with vacuum distillation............" on 6 Oct. was posted on the same day in the same
thread. Go read it and then come back if do not understand the technicalities.
gsd
|
|
Klute
International Hazard
Posts: 1378
Registered: 18-10-2006
Location: France
Member Is Offline
Mood: No Mood
|
|
I did the same mistake when I first built my aspirator, and used the same kind of pump. Thing is, these pump, 1° heat the water up significantly and
quickly, 2°, and most importantly, are rather made to have a good flow than a good pressure (HMT= 10-15m). I had crappy vacuum with such pond pumps.
I bought a cheap(er) external pump, made for watering your garden using water from a well, with a HMT of 35m.. But lower maximum flow (doesn't matter,
high enough to suit the aspirator), and result: 20mmHg with cold water, and less rapid heating of the water.
Well, after a few hours use, the wtaer heats up to 35-40°C easily, but you just need to change it evry hour or so. Using ice is a waste as it melts
very quickly and only gives you a couple torr less for a short period. Cold tap water much more practical, and gives 25-30mmHg.
There's a picture of my setup somewher eon this forum, if you want I can look up the technical data on my pump when I get the time to.
\"You can battle with a demon, you can embrace a demon; what the hell can you do with a fucking spiritual computer?\"