fume hood design

bob800 - 28-7-2010 at 17:27

I'm thinking of building a homemade fume hood. I know there's other threads on this topic, but I'm very new to fume hood design and need some basic information. Would <a href="http://www.instructables.com/id/Building-a-small-fume-hood-for-stinky-projects/">this</a> be acceptable for an amateur lab? Is the range hood ok for chemistry or do I need to buy a separate blower?

Also, what are baffles and what are they used for? Is plexiglass a suitable material for the sash? And finally what size dimensions\blower ratings are suitable for a small hood?

Any advice on building fume hoods would be greatly appreciated,
thanks

cnidocyte - 29-7-2010 at 03:48

No that fume hood isn't intended for chemistry. To make a fume hood use gypsum plasterboard for the walls since its relatively fireproof. Plexiglass can be used but it has poor solvent resistance so you may need to replace it if it gets splashed with solvent. Polycarbonate is a bit better but still isn't very resistant to non polar solvents. Ideally you could use a pane of normal glass on the inside and plexiglass/polycarbonate on the outside. The ordinary glass is highly resistant to chemicals and the plexiglass/polycarbonate is shatter resistant. If you don't want to spend too much money on it right now though plexiglass by itself is fine.

I've been wondering the same thing about baffles. What the hell are they? I looked up on wikipedia and a load of other sites but could only find vague references as to what they might be.

Picric-A - 29-7-2010 at 06:19

I believe baffles help reduce air turbulance and so it the air is drawn out of the fume cupboard quicker without any being blown back out towards the user.

That fume cupboard would be ok however the wooden walls might need replacing every so often due to spilled chemicals. If the wood is painted with an epoxy paint it should be fine.

densest - 29-7-2010 at 09:09

Flammable plastics, heat, and solvents -> splattered flaming plastic stuck to your flesh. A very bad idea. A thin piece of polycarbonate glued to window glass or between two pieces of glass is a poor person's bulletproof glass.

Range hood motors are in the air flow for cooling & to save money. Solvent vapors or corrosives will destroy the motor. The motor should be outside.

There's a cement board which is very heat resistant and quite cheap.

Discarded PC cases are a source of epoxy painted steel. It's a pain to cut up or unfold, but it is strong and usually free. "Pop" rivets (blind rivets that look like nails applied from one side) are good to stick the pieces together.

You need to change the air in the hood quickly. You need to move air into the hood faster than toxic gases fall out of the hood door. So if the door opening is 2' x 3' = 6 sq feet, moving the air 1'/sec into the hood = 360 cfm (cubic feet per minute).

The hood needs to be bigger than the biggest apparatus you plan to use in it plus enough work space to prepare any reagents plus space for a light. The smallest useful hood is 12" wide by 18" high IMnsHO.

Perhaps the fume hood discussions should be gathered together & made sticky under "reagents and apparatus" or part of "writing the textbook"?

An architect who designs chem labs wrote on a photography web site saying that the aerodynamics are critical, especially for heavier than air vapors like mercury. Gases & vapors go where they want to which is often not where you think they will.


bob800 - 29-7-2010 at 10:19

Thanks for all the help. Maybe <a href="http://developer.nicejewel.com/fumehood1.php">this</a> fume hood setup is better for a lab?

So would a blower like <a href="http://www.grainger.com/Grainger/DAYTON-PSC-Blower-1TDT2?Pid=search">this one</a> work? It's pretty expensive though, is it possible to get one less than $100 with 300 or more cfm? Will the motor last if I put it at the end of the vent? Also, what does the "@ 0.000-In. SP" mean in "CFM @ 0.000-In. : 549"?

thanks

densest - 30-7-2010 at 11:34

The jewelry maker's hood would probably be OK for HOLDING COLD INORGANIC fumes. It would be hazardous for anything heated above 50C or so and could fog or melt the door from any organic solvent other than ethanol or isopropanol.

You couldn't work in it because it doesn't have a real sash and there's no way to be sure that fumes wouldn't swirl out when the door is open.

gardenvariety - 30-7-2010 at 14:05

"@ 0.000-In. SP", in "CFM @ 0.000-In. : 549" means zero inches of static pressure. Generally you are blowing through a fan of some kind (although the math may work for compressors), and their resistance to flow, by design or dust, is measured in said inches, which I believe are vertical inches of water displaced.

My (poorly researched) rule of thumb is that 1" SP = 1/2 the cfm.

bob800 - 30-7-2010 at 17:15

Thanks. Do you think this fan : <a href="http://www.horticulturesource.com/product_info.php?products_id=6386">http://www.horticulturesource.com/product_info.php?products_id=6386</ a> would work? It's 471 cfm and it looks like the motor is separated from the air flow.

Lambda-Eyde - 30-7-2010 at 17:27

What you're after is a squirrel cage fan/blower.



bob800 - 30-7-2010 at 17:34

I thought the motor is supposed to be (ideally) separated from the fan to prevent it from being destroyed from corrosive vapors. Is that the case for these squirrel blowers? Or would they last long enough even though the motor is in the air flow?

entropy51 - 30-7-2010 at 17:36

Based on about 30 years of using a hood not too different from the link you posted, I think that hood would be good for many kinds of experiments. Just follow S.C. Wack's advice:
Quote: Originally posted by S.C. Wack  
As I've pointed out before, the original fume cupboards were powered by flame. Then ordinary fans. This was suitable for a long time.

If you're having explosions, you need to either stop pouring stuff on the floor of the box instead of in glassware, use coolant with your condenser and receiver when distilling ether or acetaldehyde, or get a bigger fan.
The trick to using a hood is not to let the vapors loose in it. Good technique, scrubbing traps, very cold water in the condenser, no ignition sources make the use of a simple hood quite safe for halogens and many organic vapors. Avoiding runaway reactions, which seem to be inexplicably common around here, is important. No need to run a 10 mole reaction unless you're manufacturing something to sell.

My hood is not too different from that one, but my hood is larger, with double plastic doors opening from each side. I also have an L-shaped baffle or "airfoil" under the doors so that a nice laminar flow of air enters and sweeps across the floor of the hood on its way to the exhaust port. I've worked with halogens, phosgene, AlCl3, Grignards, H2S and lots more without trouble.

I've yet to need to put on a gas mask, which I believe is an admission of lousy technique.

Before using a DIY hood you should test the draft with some sort of smoke generator to be sure that the air flow is adequate. Some dry ice in a beaker of water makes a pretty good smoke generator.

[Edited on 31-7-2010 by entropy51]

bob800 - 30-7-2010 at 17:45

Quote: Originally posted by entropy51  

My hood is not too different from that one, but my hood is larger, with double plastic doors opening from each side. I also have an L-shaped baffle or "airfoil" under the doors so that a nice laminar flow of air enters and sweeps across the floor of the hood on its way to the exhaust port. I've worked with halogens, phosgene, AlCl3, Grignards, H2S and lots more without trouble.


What type of blower are you using?

entropy51 - 30-7-2010 at 18:15

Quote: Originally posted by bob800  

What type of blower are you using?
It's a Dayton squirrel cage blower that looks a lot like the link you posted. Its over 20 years old and I can't read the model number, but it looks like it says 1300 CFM. It could be bigger and I will replace it with a larger one when it quits, but so far it has kept me out of the emergency room. I can make Br2 in 10 mL batches in it without catching so much as a whiff. I work on the smallest scale possible to avoid unpleasant surprises.

I wouldn't evaporate ether in an open container in a hood like this, but I have distilled ether in it many times with an ice cold condenser and receiver.

bob800 - 30-7-2010 at 18:21

So something like this would work: <a href="http://www.kalyx.com/store/proddetail.cfm/ItemID/897933/CategoryID/12000/SubCatID/2745/file.htm">http://www.kalyx.com/store/proddetail.cfm/It emID/897933/CategoryID/12000/SubCatID/2745/file.htm</a>? I don't understand why these squirrel cage blowers are so much more expensive per cfm then something like this: <a href="http://www.horticulturesource.com/product_info.php?products_id=6386">http://www.horticulturesource.com/product_info.php?products_id=6386</ a>

densest - 30-7-2010 at 18:28

Blowers like the one pictured a few posts above are perfectly suitable for a fume hood. The motor is out of the air flow. The ones bob posted immediately above will work for a while, depending on how much acid & solvent fumes they have to endure. If you're careful, they'll work for years.

Blowers differ a lot in price and construction depending on whether they have to push against resistance, such as air filters (you were going to put some charcoal or something to absorb the nasties before your neighbors breathe them?). "Back pressure" usually rated in inches of water is the key rating.

If the air flow is adequate, then the concentration of flammable material in air is below the amount needed to explode or sustain combustion. That's why the flame-powered hoods worked.

And as S.C.Wack & entropy51 have pointed out, a hood is not an excuse for sloppy setup. It's a life preserver (or eye/lung/hand preserver). Some preparations require moving fuming beakers around. That's fine, and that's what a fume hood is for - just put the fuming mess into something which captures and disposes of the fumes ASAP. That's how I handle aqua regia and other lung-unfriendly things.


[Edited on 31-7-2010 by densest]

entropy51 - 30-7-2010 at 18:29

Quote: Originally posted by bob800  
So something like this would work: <a href="http://www.kalyx.com/store/proddetail.cfm/ItemID/897933/CategoryID/12000/SubCatID/2745/file.htm">http://www.kalyx.com/store/proddetail.cfm/It emID/897933/CategoryID/12000/SubCatID/2745/file.htm</a>? I don't understand why these squirrel cage blowers are so much more expensive per cfm then something like this: <a href="http://www.horticulturesource.com/product_info.php?products_id=6386">http://www.horticulturesource.com/product_info.php?products_id=6386</ a>
$100 is not expensive for a blower, but it looks like the motor is in the gas stream, which is a no no. Nobody said chemistry was an inexpensive hobby. A good blower will last a long time if you scrub your waste gas stream before it goes into the blower.

densest - 30-7-2010 at 18:36

@bob800 - the blower you referenced will not push against an air filter. If you release corrosive or solvent vapors, it will degrade and eventually fail. Murphy's law guarantees it will be at the most disastrous time.

I use a similar fan to vent my glass working area. For that, it's fine - there are a small quantity of nitrogen oxides which would be very bad for my lungs but are not sufficient to corrode the fan motor. I would not use it to remove acid fumes.

bob800 - 30-7-2010 at 18:37

Thanks so much for all the help. So if my hood is 2' wide by 1.5' high by 14 inches back then would this blower work?: <a href="http://www.kalyx.com/store/proddetail.cfm/ItemID/897933/CategoryID/12000/SubCatID/2745/file.htm">http://www.kalyx.com/store/proddetail.cfm/It emID/897933/CategoryID/12000/SubCatID/2745/file.htm</a> EDIT: Oops, that's the one you said was in the gas flow. How about the 400 cfm version of this one: <a href="http://www.planetnatural.com/site/search_engine.html?mv_session_id=j4Fisob8&search_clear=1&criteria=active+air+blowers&x=0&y=0"& gt;http://www.planetnatural.com/site/search_engine.html?mv_session_id=j4Fisob8&search_clear=1&criteria=active+air+blowers&x=0&y=0</ a>

[Edited on 31-7-2010 by bob800]

[Edited on 31-7-2010 by bob800]

densest - 30-7-2010 at 18:46

I wouldn't use it - the motor is inside the air stream. I've had to replace such a fan twice in the hood over my stove.

Something like 1963k37 from http://www.mcmaster.com is good. It's about $400, which means you could probably find one surplus for $100 or less.

Don't skimp on your safety!!!!!!!!!!!! I'm still regretting a very "minor" injury to one finger. It took a year to recover from a day of too much nitrogen oxides.

If you want to do chemistry on a small budget, you have to scrounge good equipment from unlikely places. There are a lot of surplus equipment dealers around the planet. Your job is to know better than they do what something is worth or to know how to use something that they have better than anyone else local.

[Edited on 31-7-2010 by densest]

bob800 - 30-7-2010 at 18:47

Sorry, I posted the wrong blower. See my edited post.

densest - 30-7-2010 at 18:50

That one looks good. Do the math for your hood - I use an arbitrary value of 1 foot per second of air into the hood through the door.

bob800 - 30-7-2010 at 18:53

What's the formula for calculating that?

densest - 30-7-2010 at 19:05

I gave it above - cfm of fan / ( 60 * area of opening in square feet)

bob800 - 30-7-2010 at 19:08

Thanks - sorry, I didn't realize it was the same thing

bob800 - 30-7-2010 at 19:12

OK- So if my opening is 1.5 feet high, 2 feet long and 465 cfm then it would be 2.66 seconds? Is that correct?

[Edited on 31-7-2010 by bob800]

thorazine - 30-7-2010 at 19:59

I was wondering if was a good ideia use parquet on the floor of the fumewood! It's chemical resistant and cheap..

entropy51 - 31-7-2010 at 06:52

Quote: Originally posted by thorazine  
I was wondering if was a good ideia use parquet on the floor of the fumewood! It's chemical resistant and cheap..
I use ceramic floor tiles. After over 10 years use they are just like new, but as S.C. Wack suggests, I don't pour chemicals on them. I try to always pour chemicals over a metal or plastic tray that will contain a spill.

bob800 - 31-7-2010 at 07:24

Also, what type of ducting should I use? Is ordinary galvanized steel ducting OK? Also, how necessary is it to place the motor at the end of the duct if it's separated from the air flow?

[Edited on 31-7-2010 by bob800]

watson.fawkes - 31-7-2010 at 12:46

Quote: Originally posted by bob800  
Also, how necessary is it to place the motor at the end of the duct if it's separated from the air flow?
Ducting is rarely perfectly airtight without using the proper duct mastic. Before the fan, duct generally has negative &Delta;P with respect to ambient; after it, positive. Be careful about that leakage flow after the fan.

MagicJigPipe - 31-7-2010 at 17:14

What about glass over hardwood for the "floor"? What is generally used in commercial fume hoods? In the fume hoods at my school they seem to be some sort of very dense painted (black) wood. What is this material? The sinks are made of the same material, by the way.

bob800 - 2-8-2010 at 09:21

The fan I'm considering supplies this chart:

CFM @ 0.000-In. SP 463
CFM @ 0.100-In. SP 425
CFM @ 0.200-In. SP 388
CFM @ 0.300-In. SP 338
CFM @ 0.400-In. SP 250
CFM @ 0.500-In. SP 161
CFM @ 0.600-In. SP 50

Is this drop acceptable for a fume hood blower? How much resistance does an average duct have? I measured my hypothetical duct to be roughly 45" inches total, which includes 1 turn. The blower is a Dayton 1TDR9, here the Grainger link: <a href="http://www.grainger.com/Grainger/wwg/search.shtml?searchQuery=dayton+blower+1TDR9&op=search&Ntt=dayton+blower+1TDR9&N=0&sst=subs et">http://www.grainger.com/Grainger/wwg/search.shtml?searchQuery=dayton+blower+1TDR9&op=search&Ntt=dayton+blower+1TDR9&N=0&sst=sub set </a>

[Edited on 2-8-2010 by bob800]

peach - 2-8-2010 at 09:56

You could line the duct with plastic to improve it's chemical resistance and sealing.

For the surface, I'd cover it with a sheet of toughened glass. Alternatively, buy an old, shallow, wide Belfast sink. I don't know if they're called that in the US. The big, heavy, white ceramic ones you'll often see farms being retired to water or feed troughs. You might call them Shaker sinks in the US, I'm not sure.

They're glazed ceramic, needless to say, and very heavy duty. They also come with a premoulded 'trough' shape and a plug hole which is excellent, because it means any spills will slosh up the sides and back down, then down the plug, which can be connected to a bin of some sorts. Also, if a fire breaks out, it's not going to spill out of the sharp corners or melt through seams at those corners.

Additionally, the glaze is a continuous surface, there are zero joins. Which means there's nowhere for things to seep or rot through. It's also mirror smooth, making it hard for spills to stick.

All you'd have to do then is build up the walls. 1" thick play would be very stable. You could then add a lining of plasterboard, which has a 30 minute fire rating for a half inch sheet. Then tile over that and you'll have a nearly 100% ceramic lined hood. Silicon up the seams with high temperature silicon.

To save effort and time, get the timber yard to chop the 1" ply to size on their wall saw, for a few dollars; measure three times, carefully first. The plasterboard can be glued to the ply. You can cut it to size by scoring it along a a straight edge (bit of timber) and then snapping it, works perfectly and is 'the plaster's preferred method'. The sizes here don't need to be mm perfect, you can have 1/4"+ gaps around the edges and be okay. Buy some cheapo tile adhesive and some dirty cheap white, flat bathroom tiles, stick them on. Grout. Bingo...

Commercial hoods tend to use high density plastics or stainless. But the ceramic will fare better still; for example, the corrosive gases and acids will rot the finish of stainless. Normal stainless also has a 'grain' to the surface, allowing things to get trapped in those. The chemical resistance, safety properties and complexity, depend how handy you are with your handies, as always.

[EDIT] CAN WE PLEASE MERGE AND STICKY THE FUME HOOD QUESTIONS? YOU GUYS WILL NAIL REPEAT QUESTIONS IN THE ASS, LET ALONE THOSE COMING FROM NEW THREADS, AND YET THIS ONE IS HERE ALL THE TIME. IT'S DILUTING THE INFORMATION AND CLOGGING THE FORUM. STICKY IT. I'D ALSO RECOMMEND A STICKY FOR VACUUM AND HOTPLATE QUESTIONS. [/all caps]

[Edited on 2-8-2010 by peach]

watson.fawkes - 2-8-2010 at 09:58

Quote: Originally posted by bob800  
How much resistance does an average duct have? I measured my hypothetical duct to be roughly 45" inches total, with 1 turn.
Search for "duct loss" to find many resources, such as this nomogram. Such a short run of duct, though, has minimal loss. Much more important is going to be how your intake air is supplied. If your hood is in a relatively-sealed room with closed door, you'll have difficulty sucking in air, and total system resistance will be high. If you provide an explicit inlet to the room for makeup air, you'll have little problem. Often makeup air is supplied as an air curtain in front of the hood. There's no need for a small system to have the makeup air supply be powered.

bob800 - 2-8-2010 at 18:05

Thanks for all the help. I found a bunch more fume hood threads which answered my remaining questions (I didn't see the 'subject only' button in search. Duh! 5 billion more threads).

However, I still have a remaining question: How much space does a 24/40 ground glass distillation setup take? I don't have one (I currently use a retort for corrosives) but may buy one eventually.

Quote: Originally posted by peach  

[EDIT] CAN WE PLEASE MERGE AND STICKY THE FUME HOOD QUESTIONS?


Here, I made a url for every single fume hood thread (don't know how to make a sticky thread, so someone else can copy/paste this into it):

[EDIT] Make sure you have each link on only 1 line. A few links don't fit and won't work unless you delete the spaces in the middle

Code:
[url=http://www.sciencemadness.org/talk/viewthread.php?tid=14233&goto=search&pid=183724]fume hood design[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=14158&goto=search&pid=182505]fume hood Vs. Range hood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=13308&goto=search&pid=170307]Questions about fume hoods and Filtering Exhaust[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=13246&goto=search&pid=169350]Some fumehood trivia[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=13238&goto=search&pid=169251]Using a chimney as an exhaust route for fumehood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=13052&goto=search&pid=166104]DIY fume hood - how many air exchanges per minute?[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=12400&goto=search&pid=155515]idea for slight improvement of fume hood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=11841&goto=search&pid=147485]How to Make a fume hood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=11755&goto=search&pid=145803]fan selection for homemade fumehood - help[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=11754&goto=search&pid=145799]what can i coat the inside of my homemade fumehood with?[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=11145&goto=search&pid=135726]homebuilt fume hood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=10425&goto=search&pid=124994]motor for fumehood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=9058&goto=search&pid=104431]Installation of fume hood[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=8949&goto=search&pid=102993]DIY fume hood air flow design optimization[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=7253&goto=search&pid=82753]fume hood Construction[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=5260&goto=search&pid=59587]Hows this for a fume hood?[/url] [url=http://www.sciencemadness.org/talk/viewthread.php?tid=3903&goto=search&pid=43862]fume hood safety[/url]


Hope this helps.








[Edited on 3-8-2010 by bob800]

peach - 2-8-2010 at 20:56

Quote: Originally posted by bob800  

However, I still have a remaining question: How much space does a 24/40 ground glass distillation setup take? I don't have one (I currently use a retort for corrosives) but may buy one eventually.


I worked it roughly out in my head but thought the numbers were off, so I quickly set my B24/29 to check. And was much pleased to discover I was within a few cm of the rough value. It's about 70cm tall if you're using a hotplate, 1l RBF, three way head, thermometer adapter and thermometer.

I'm using the shortest length coil condenser Quickfit do, which is something like 26cm in length from memory. The longer version is only slightly longer and the short one condenses all the solvents I've put through it within the first turn or two of the coil.

In total, its about 70cm wide again.

Of special note is that it will actually get wider if you use smaller flasks, who's addition funnels usually stick out the side at an angle as opposed to going in straight up (as they do for 1l and up flasks).

The height will go up by around 36cm if you add a standard length Vigreux column.

You'll need an extra 10cm to clear tapers as you put it together, exchange things, take it apart...

And a decent amount of room on either side of the glass for adding hoses. In particular, vacuum hose is extremely thick and has quite a large bend radius, so you'd want to make sure the vacuum enters the hood on the same side as you like having your receiver. I always have it set up left to right. Not sure, but I expect 'the lefties' may do it the other way round (unless forced not to in school due to the layout of the services on the fume hood).

Good work on the link collecting. Now we need A MODERATOR to merge them and sticky.

[Edited on 3-8-2010 by peach]

bob800 - 3-8-2010 at 13:49

Thanks for measuring your setup. OK, I've figured out almost everything, however I need some feedback:

1. Dimensions: You said your setup is 70 x 70 cm, so I'll make my hood 2.5 x 2.5 feet (76 cm). If I'm correct, 2.5 feet * 2.5 feet * 60 = 375 cfm, which is in range of my blower. I don't understand why the depth of the hood isn't taken into the calculations? Should it be? PLEASE CORRECT ME IF I'M WRONG

2. Blower. I think I'm getting the Dayton 1TDT2. It's $164 straight from grainger, but I found a few $99 new on eBay. It's 549 cfm at 0.00 In. SP. Here's the full chart:

CFM @ 0.000-In. SP 549
CFM @ 0.100-In. SP 538
CFM @ 0.200-In. SP 510
CFM @ 0.300-In. SP 500
CFM @ 0.400-In. SP 480
CFM @ 0.500-In. SP 450
CFM @ 0.600-In. SP 435
CFM @ 0.700-In. SP 390
CFM @ 0.800-In. SP 360
.

Is this OK if I only have 45" inches of ducting including 1 turn as long as I make sure air can get into the sash?


3. Lining-Material- Does epoxy paint\spray paint stick to wood? If so, that's what I'm going to use, maybe ceramic tiles if it's not resistant enough.

4. Sash- I know it's not perfect, but I don't know what else to use other than regular glass with a sheet of polycarbonate glued to it. I'm not going to be mixing up nitroglycerin or anytingin my hood, so I'm not too worried about explosions. I'll use the bolt idea stolen from some other members of science madness to provide a few fixed settings.

5. Baffles- I don't think I'll install baffles unless my hood doesn't work properly. Do you think they're necessary?

6. Ducting- PVC seems too hard to work with, and I can't find anything wrong with plain galvanized steel ducts as long as I don't boil HCl for hours on end.

Please post any feedback! Thanks.

entropy51 - 3-8-2010 at 14:11

Quote: Originally posted by bob800  
Thanks for measuring your setup. OK, I've figured out almost everything, however I need some feedback:

1. Dimensions: You said your setup is 70 x 70 cm, so I'll make my hood 2.5 x 2.5 feet (76 cm). If I'm correct, 2.5 feet * 2.5 feet * 60 = 375 cfm, which is in range of my blower. I don't understand why the depth of the hood isn't taken into the calculations? Should it be? PLEASE CORRECT ME IF I'M WRONG
You're wrong.

The equation is Face Velocity x flow area = CFM . 60 is the conversion for minutes to seconds. If you wanted 75 ft/min face velocity then the required CFM = 75 x 2.5 x 2.5 = 469 cubic ft per minute. If you mostly operate with the sash half closed, then the flow area and the CFM is also cut in half for the same face velocity and so you need about 235 CFM.

But you will need to use the charts that are discussed above to find the static pressure drop in your duct to find the actual CFM the blower will pull when attached to your setup.

bob800 - 3-8-2010 at 14:48

If I read correctly the nomogram for 5" and 600 cfm shows 8 INCHES FRICTION!!!.:( That's way too much... Could I adapt the pipe to say 8" and read the chart from there? Or would the adapter still add a lot of friction? Or is the motor wheel just not physically big enough?

[Edited on 3-8-2010 by bob800]

[Edited on 3-8-2010 by bob800]

peach - 3-8-2010 at 15:05

Entropy is correct, it's the area you need to consider when working out the flow velocity.

What's wrong with using PVC? There's a ton of it at the hardware store, you can cut it with a hacksaw and the fittings simply push together to provide a seal. The seal also can't fall apart, as the washers point inwards, like wipers, which prevents the pipe from sliding back out. In fact, it's physically impossible to pull them apart without releasing the collet. That's a lie, I did manage to pull a connection apart at one point, using the bucket of an excavator that was being driven by ten HP+ of hydraulic pumping; I did so intentionally before anyone jokes about it :P

110mm foul drain pipe isn't very expensive, it has a large ID, it's very smooth and fairly resistant. If you use 110mm, taper the edges of the cuts with a file and lubricate them up with something slippery; washing up liquid, KY, Molykote, Krytox, vacuum grease, Teflon grease, Vasoline, anal lube....



The lubricant won't make them fall apart, but they will slide together a lot quicker. Awwwww....

If you're concerned about safety, use toughened glass. If you call around the custom glass places, they'll have it. Bathroom suppliers or people involved with things like glazing for schools are good bets. The bathroom places will use it to produce things like panes for showers, or curve it to make walk in showers. You can even buy the toughened panes as components for showers from numerous places. But you'll probably need someone with a diamond saw to cut it, as the scratch and snap method is unlikely to work.

If you use a sheet of polycarbonate, that's fairly expensive to start with. Not as expensive as the toughened glass is likely to be, but on the way. Then add on the complexity of making a dual layer screen and the potentially shit performance it may give; e.g. bubbles between the layers, delamination, build time, etc. And you're on the way to the toughened glass being a time, money & effort saver. Delamination is a good probability if it's being exposed to odd, concentrated solvents, acids, bases and other reactive thingymejiggies.

If you choose the polycarbonate way, you'd probably want the plastic on the outside, as the glass will be more resistant. Whereas the plastic may fog up when exposed to certain things.

If you're doing things involving highly corrosive or toxic gases, the fume hood should be there only to mop up the escapees, not the bulk of it. Using greased tapers, keck clips and wash bottles will take care of 99%+ of the fumes. I can easily generate large quantities of HCl(g) and not even be able to smell it with my nose against the glass.

Fume hoods encourage lazy and dangerous attitudes towards letting things escape and hoping they'll be picked up and absorbed or blown off. Get rid of the risk as it leaves the glass.

It's akin to HazMat suits. The only time you should need one of them is when you're cleaning up someone else's mess or an accident, not for everyday work (unless you work in a nuclear power station, germ warfare or nerve agent factory, and even then, it shouldn't be in the atmosphere in the first place).

[Edited on 3-8-2010 by peach]

entropy51 - 3-8-2010 at 15:11

The face velocity you need depends on the toxicity of the material you are working with. Unless you plan to work with HCN or phosgene you may need less face velocity during most of the operations you conduct in the hood. The required face velocity also depends on the amount of toxic or stinky stuff you plan to release into the hood. If you plan mostly small scale experiments you can also get by with a lower face velocity.

My hood has a very low face velocity with the doors ("sash") open. But I keep the doors closed during most of a reaction or distillation and the face velocity is much higher during those conditions. It apparently works because I am here to talk about it in spite of having done some risky work in that hood.

I would recommend that you stick close to the design space (CFM, duct dimensions, face area) that Magpie has used in his hood because he is a chemical engineer and he uses his hood quite a bit.

I'll say it once more: keep the vapors and fumes confined in the glassware and scrub the effluent stream leaving the glassware. It's not a good idea to boil HCl or evaporate ether in open vessels in a home-made hood.

If you don't conduct high hazard operations in your hood, I suspect the blower you mentioned last is not an unreasonable choice. If need be you can adapt the scale of your reactions to the performance of the hood.

Magpie may have had similar experiences as I did: when I took organic there was one hood in the entire lab with 30 students. We poured the Br2 into our addition funnel in the hood and then took it back to the bench (with a stopper) and attached it to the flask and conducted the bromination. It was unpleasant, but no one died. Moderately good ventilation is better than none at all!

[Edited on 3-8-2010 by entropy51]

peach - 3-8-2010 at 15:17

Quote: Originally posted by entropy51  
It apparently works because I am here to talk about it in spite of having done some risky work in that hood.


:D

Not sure about that, I dread to think what's lurking in my balls to surprise me later in life. I should probably make better use of them now before they go supernova, then shrivel up and need chopping off.

If I were working with anything you mentioned, I'd have gotten rid of it before it ever left the glass. E.g. base washes, solvent scrubbers and so on... back to the wash heads, my best friends in the ethereal glassware domain.

This is like I was saying, I think students need teaching that kind of thing first. Letting them get straight to the fume hood encourages the 'I'll just run this lethal stuff in a beaker' mentality. Like the engineers who can CAD model a bridge but have never picked up a hammer.

watson.fawkes - 3-8-2010 at 15:24

Quote: Originally posted by bob800  
If I read correctly the nomogram for 5" and 600 cfm shows 8 INCHES FRICTION!!!.:( That's way too much... Could I adapt the pipe to say 8" and read the chart from there? Or would the adapter still add a lot of friction? Or is the motor wheel just not physically big enough?
That's 8" water loss per 100' of duct. If you have 50' of duct, that's only 4" of loss.

bob800 - 3-8-2010 at 15:30

Quote: Originally posted by entropy51  

I would recommend that you stick close to the design space (CFM, duct dimensions, face area) that Magpie has used in his hood because he is a chemical engineer and he uses his hood quite a bit.


I agree except that Magpie's hood is much bigger than I need (and the motor is too expensive for me-$400 new). All I need is to be able to barely fit a distil. setup.

BTW, do you know of any websites to get surplus blowers? eBay doesn't have a great selection...


[Edited on 3-8-2010 by bob800]

[Edited on 3-8-2010 by bob800]

bob800 - 3-8-2010 at 15:34

Quote: Originally posted by watson.fawkes  
That's 8" water loss per 100' of duct. If you have 50' of duct, that's only 4" of loss.


Oh..... Why didn't I see that? That changes everything...

entropy51 - 3-8-2010 at 15:43

Quote: Originally posted by bob800  
I agree except that Magpie's hood is much bigger than I need (and the motor is too expensive for me-$400 new). All I need is to be able to barely fit a distil. setup.
I think that blower may work for you. Just try to keep your ducts as short as possible and avoid bends. There is a certain amount of trial and error involved in any engineering project. And as you've learned, financial considerations involve compromises in performance. Your first trial at a project like this won't be perfect, but the perfect is the enemy of the good enough. I was taught to design to as many significant digits as possible, but that the final delicate adjustment usually involves a ball-peen hammer and a crowbar.

bob800 - 3-8-2010 at 15:44

So if my duct is around 45 INCHES long, at 5" diameter at 549 cfm, then I'd only have ~ 0.3 inches of S.P.! Then my blower would be PLENTY strong enough! Am I correct?

bob800 - 3-8-2010 at 15:55

Quote: Originally posted by peach  

What's wrong with using PVC? There's a ton of it at the hardware store, you can cut it with a hacksaw and the fittings simply push together to provide a seal


True. But what's wrong with ordinary galvanized steel ducts?

watson.fawkes - 3-8-2010 at 16:36

Quote: Originally posted by bob800  
So if my duct is around 45 INCHES long, at 5" diameter at 549 cfm, then I'd only have ~ 0.3 inches of S.P.! Then my blower would be PLENTY strong enough! Am I correct?
It's true if that's 45 inches of straight duct without bends. Bends add to static pressure drop.

entropy51 - 3-8-2010 at 17:00

You will also have static pressure drops in the adapters you need to connect your hood to the blower and your blower to your exhaust duct. The blower outlet is square and your duct is probably circular. These losses will likely exceed that of the friction loss in the exhaust duct. But I still think you are in the ball park on the blower.

A smooth steel duct will be fine if you don't let a lot of corrosive fumes enter it. If you do let corrosives into it, the blower will also not last long.

bob800 - 3-8-2010 at 17:03

Quote: Originally posted by watson.fawkes  
Quote: Originally posted by bob800  
So if my duct is around 45 INCHES long, at 5" diameter at 549 cfm, then I'd only have ~ 0.3 inches of S.P.! Then my blower would be PLENTY strong enough! Am I correct?
It's true if that's 45 inches of straight duct without bends. Bends add to static pressure drop.


There would be 1 bend which I included in the 45". I found another graph on engineeringtoolbox for duct bends (haven't read it yet): http://www.engineeringtoolbox.com/air-duct-minor-loss-diagra...

[EDIT] @entropy51: Yes, the outlet is square. There's a window directly above my work-area that I'm going to take out and replace with a metal "window" with a hole cut. I think I'm going to mount the blower directly into the metal sheet so I don't need a bunch of custom-made adapters.

[Edited on 4-8-2010 by bob800]

[Edited on 4-8-2010 by bob800]

entropy51 - 3-8-2010 at 17:21

If the window is visible to passers by, it might be better to use a thick sheet of clear Plexiglas (PMMA) over the window and attach the blower to that. That's what I've done and it's never attracted any attention from the curious folk who seem to be everywhere. Even better to also plant a shrub in front of the window. Even better if you have no neighbors, but a low profile is always good.

bob800 - 3-8-2010 at 17:24

Quote: Originally posted by entropy51  
If the window is visible to passers by, it might be better to use a thick sheet of clear Plexiglas (PMMA) over the window and attach the blower to that. That's what I've done and it's never attracted any attention from the curious folk who seem to be everywhere. Even better to also plant a shrub in front of the window. Even better if you have no neighbors, but a low profile is always good.


Don't worry, my duct will be covered by a bunch of bushes and would look like any other laundry dryer duct if someone saw it.

peach - 5-8-2010 at 08:23

Are you filtering the output?

If not, the legality of what you're making aside, those that be may become upset should someone ever notice the odd smells and their skin tingling as they pass by; if they can get close to the vent. Some of the foul smelling things I've produced have released copious clouds of persistent fumes for hours, that people can smell a long, long way away. Even after being blown out, they'll hang in the air like smoke from a fire; causing people to call the authorities out thinking your house is on fire.

High up on the roof is the best place for it, where it'll be diluted and washed away by the higher air stream. There is a pocket of much slower moving air around ground level.

Failing that, having it vent to the back garden side of the house will give it time to spread out a little more.

It's lazy, and dangerous, to be letting the harmful fumes escape the glass in the first place. Morally, I think it's a bit of a shitty thing to be then sucking the fumes away from yourself and spraying them towards someone else. Particularly if there are kids around.

Activated carbon filters are supplied in a vast array of sizes and capacities by the hydroponic stores selling to cannabis growers. They work exceptionally well for a lot of odors. I've run them and actually had the place smell fresher than it did before I started producing things that stink.

Sorry if this is redundant and you will be filtering. I just don't see the point of going on about safety and then risking someone else's when it involves hammering together a big hood, all this work and when the filter isn't much more.

All thar best,
John

bob800 - 5-8-2010 at 09:12

My only option is to vent at ground level (with a bunch of bushes in front), but I'll follow your advice and neutralize as many fumes as I can before they can leave the glass, and work on a small scale. And I can always work on windy days if needed.

If it still stinks outside, I'll consider adding a filter as you suggested. How much resistance do those carbon filters add? Do you think my blower can push the air through one and still have enough face velocity?

[EDIT] Maybe I could just blow the fumes through a good filter and release them back into my lab. Then I wouldn't have to worry about getting the duct through the window and fumes stinking up the whole neighborhood etc. etc. AND having a source of air to be sucked into the hood wouldn't be a problem. How many gasses do those filters stop? I'd need them to stop: Cl2, NO2, SO2, Br (small amounts) small amount of acid fumes from conc. solutions, burning fumes like burning Mg or steel wool, etc. etc. I'm not going to be doing much organic chem, so whatever scary fumes are created through O-Chem are not a problem. What do you think about a ductless hood?


@peach: Thanks for the suggestions, you're not being redundant.

BTW, I've heard of "gas wash heads", but what are they? How are they different than gas washing bottles like this?

[Edited on 5-8-2010 by bob800]

peach - 5-8-2010 at 15:43

If you're the kind of odd ball person I find floating around a number of these forums, you could also run the work once it's gone 11 or 12 at night, when most people will be home and probably in bed. I'm quite often up at 4am messing around. Watching people walking up and down the road hitting things with a garden spade, for example. :P

If you blown enough corrosive or dirty shit out the exhaust, that bush in front of it may begin to take on a chemically toasted appearance. If the rest of the garden is vibrant green, the bush may make the exhaust stand out even better to passers by than the exit it's self, when they hear the fan noise and notice a bush near were it's coming from has gone brown.

You may wish to add a one way flap to the outlet (like a piece of rubber sheeting), such that a gust of wind isn't going to blow the fumes back through the hood.

Running it on wet or rainy days will also knock a lot of things out of the air. HCl(g) will fall out quickly, for example. You can easily implement a method to force a constant rainy day on the output by simply sticking a tiny fogger or misting nozzle towards the end of the outlet.

In the UK, we have a brand called Hozelok that sell a ton of cute little micro-irrigation things at the garden centres. They sell a number of nozzles for producing a cone or fan of mist for the plants. All you need do is get one of those and hook it up to a tap, then turn it on when something nasty starts coming off. The water usage rate will be significantly lower than it would for an aspirator in the sink. If you're using PVC, corrosion won't be an issue and you can simply drill a hole for the nozzle to poke in and silicon or epoxy it in place, or a few of them at various angles down the length to get a good scrub action on the go.

I think I actually have some of those nozzles in the garage. If I can find them, I'll post you one or two for free.

This is a method used in some commercial fume hoods. To deviate a little for a history lesson, in 'ye olde days' they didn't even use a fan. They'd have a flame at the back of the hood to draw the fumes out by convection. This also happens if the exit is high up, it's called the chimney effect and it's partly why chemical plants and power stations use chimneys, the air rushing over the top creates a venturi like effect and sucks gas up automatically. The helical swirls you see on some are vortex breakers, so the airflow around the top doesn't place too much bending load on the stack. U-Boats used this trick in water when they realized in WW2 that raising the periscope effectively draws a long white line of foam, created by turbulence, towards the exact location of the sub, for a plane overhead to line up with and then bomb. :D Hoffman also commented that he was using flame aspirated hoods in his lab when he produced LSD, when discussing how basic their equipment was back then despite dealing with something so powerful. For biologists in particular, the flame has the additional benefit of killing anything organic trying to escape. State of the art aerospace design also makes use of vortex properties. Those little fins you see sticking vertically out of the wings on an airliner are vortex catchers. By stabilizing the vortex under the wing, it creates high lift efficiency. If they're allowed to burst under the wing, the pressure difference drops. And the randomized popping creates turbulence. The same effect is now being proposed for wind turbines, to increase their efficiency. And, yet again, it's also seen in nature. The bumps on the fins of big whales have evolved to be there as they allow the fin to cut through the water better by controlling the vortexes flowing over it, and so, the turbulence.

The activated filters are also standard for modern commercial hoods, the recirculating hoods are largely just a carbon filter. Which worries me somewhat given the variability of things that'll go through them, as anyone who's looked at respirators will know one cartridge will not safely absorb every fume. E.g. ammonia and monoxide will pass through carbon filters okay.

I don't know about the precise velocities and such that you'll achieve with the filter in place, but you can buy huge ones and you can always modify the fan selection. If commercial hoods use them, you'll be able to do it, it's just a matter of cost. Anything another human has designed or built, you can understand or replicate given enough effort.

Another option would be to use a HEPA filter with an activated carbon back, which will also allow a lot of gas through. You could even fit the filter behind the baffles, such that most of the nasty shit is absorbed within the hood before it gets to the duct. Big, high flow rate, low pressure drop HEPA filters will probably work out even cheaper than the hydroponics scrubbers, as they are manufactured in vast quantities, globally and are very basic in terms of construction, but also effective enough for use in silicon wafer labs, hospitals and biological laboratories. I've never taken apart a commercial recirculating hood. But I wouldn't be surprised to find a HEPA filter in a number of them.

You're looking for active carbon HEPAs, and the flat kind for the lowest price. You only need the filter element (which are available as consumables for replacement when the originals run off their ratings or start to clog, after about a year to three of heavy use). To drop the back pressure and improve the flow rates even more, you could then pull off the micronic filter stage and just have the activated carbon. Taking care not to pull off whatever is being used to keep it in place, or you'll look like a cartoon character that's just had a bomb go off in their face when you turn it on (the carbon is incredibly fine dust and black, so there'll be something inside containing it).

All the premade filters aside, you could even produce your own, spend even less and likely have it work even better than a commercial option. I expect it'd take less than an hour to put together, probably just a few minutes, as the parts are designed to fit together in seconds with no tools.

To do so, I'd;

1.) Buy a bag of activated carbon powder.
2.) Try to find out what it's rough particle size is
3.) Find something with a mesh rating that's slightly smaller than that (like woven cloth or wire mesh); some places sell wire mesh so finely woven it will filter micronic moisture droplets out of the air, and they often sell it in various metals (like stainless) and in little 1 foot swatches.
4.) Get some PVC sewer pipe and fittings.
5.) Cut a length of it and stick a swatch of material over it, then press the fitting on
6.) Pour in the carbon
7.) Repeat stage 5 for the other end.

That requires zero input from someone who needs the capital to set up a factory and machines, pay peoples wages and make money on it. It's also balls easy. If you make the carbon section a bit bigger than the charge you add, the particles will also get some opportunity to move around, decreasing the build up of concentrated layers. It'll also mean the filter has less opportunity to pack and produce back pressure.

In terms of the mesh, the wire will distort the least (messing with the pore rating) as you jam it in place. But more active metals may also rot open, so a plastic may be better if you're not getting stainless. The pore size should be a little below the dust particle size, but not a lot more. You just want to trap the dust, making it any smaller will make it more expensive and produce more back pressure, making the fan more expensive too.

The American Society for Testing and Materials (ASTM) counts activated carbon as powered (PAC) when the mesh goes above 80, or below about 0.17mm for us newbs who can't visualize mesh numbers. This is the same stuff used in the expensive commercial filters.

The stainless mesh is expensive. A 100 micron (0.1) teflon coated mesh is about £20 per square foot. 100 mesh copper (won't react with HCl(g)) is about £10. Aluminium is around £10. Bronze is £6.25. Brass is about £5. Using a plastic or fabric mesh will be even cheaper.

The prices are from here, but search around and you'll quickly find others doing them by the square foot and they may have discounts on the meshes you're after in stainless

Have a look at this bit as well, where they discuss the different weave patterns. Dutch is the toughest. Sintered won't distort but costs a lot more.

It's likely that you'll want to use the largest particle size available (or even a small grain size) if you pack a good length of the pipe, as it'll be a layer much thicker than most filters so the super fine dust will produce more back pressure. Using larger granulated material will also make the mesh to retain it even cheaper. I would recommend against packing something like the following access fitting full of fine power; enough to create a layer will do.

Anywhere from 6 months to years later, depending on what goes through it and how often, it'll need recharging. Pulling apart standard 110mm fittings is hard work or essentially impossible. You'd need to pay a little more for ones with collets that unscrew. A far, far, far, farrrrr easier option is to contruct your carbon chamber from a 'drain access', 'drain entry', 'rodding' port or 'access pipe'. That's a short length of pipe with an unscrewable port on the side. Do the same with it as above, but now when it comes to recharging you can simply unscrew the port, pour out or hoover up the carbon, pour some fresh stuff in and go.

You can also buy sewer pipe in 160mm, but it's for commercial uses (so less likely to be found in regular hardware stores) and is more expensive (for obvious reasons). There is an added bonus to using what is called 'vent soil pipe' (the black or grey stuff for above terra firma usage), in that you can make it look like a normal foul drain even if it's sticking out the side of the house by simply running it to ground level (although I've heard you crazies in the US run the foul pipes inside the walls sometimes, they're all outside here because all the walls are solid brickwork, making it hard to do the wall trick). Also, have you considered hooking the hood to the actual sewer? That way there's zero chance anyone is going to connect the smell or noise with your house. Especially so if the nearest grid in the street isn't directly outside your drive (as it is for me).

In the UK, the foul drain that leads out to the main pipe in the road is connected to it, by law, via a water trap, to prevent potentially dangerous pathogens and such flowing back up the pipe from material rotting in the public sewerage system. We don't put the traps there, they're put in by the council when they lay the main sewers and hook the houses up in the road. There is also a 'vent pipe' which runs all the way up the side of the house to the roof, where any nasty smells or germs can wash away and air can get in to keep the pressures in the pipe correct for the traps to function.

Hence, if I was to blow a hood's exhaust down my sewer, it'd blow out two stories up where no one would smell it, feel it on them or notice the noise. Even better, if the vent pipe is near the chimney, to a passing glance any persistent fuming is likely to written off as the result of a log or coal fire venting where it's supposed to. With the exit being high up, you stand a better chance of benefiting from a natural aspiration, chimney effect too. Ever seen the water in your toilet being sucked back when it's windy out? That's the aspiration effect, producing quite a fair bit of vacuum in the pipe (the effect doesn't work very much in reverse, blowing things back down it). And you have somewhere for the water from your misting nozzles to drain to. Win, win, win eh? Bar the possible need for a stronger fan (or you could get used to keeping the slash closed over, and keeping in mind that people on here go on about using baffles and adjustable fans to purposefully turn their hoods down when the sash is part closed). Someone has already built a roof mounted, water trapped exhaust for yar. Yars just need to be connectin' to her.

To pick up on the baffling point, I'm not sure how important it is or if people know about it, but baffling a number of fans causes the blades to begin turbulating and stalling and the flow numbers aren't the same as simply turning down the juice into it; the rotor isn't slowing, it's slipping through the air rather than catching it, like a cavitating propeller. Centrifugal fans are prone to this effect. When it starts happening, the flow rate doesn't fall in a predictable fashion, it vertically drops off. As was the case for the U2 spy plane, who's wings would snap off if it sped up too much or stall if it slowed down too much at it's 70 - 80k feet cruise altitude; the margin for error in the speed was 10 knots! :o This is called the 'death corner' of a planes aerodynamic performance. Airliners suffer a similar problem, but not as severely. Providing another example where an electronically controlled fan speed will probably be a safer bet than doing it mechanically, as the stalling won't occur when you turn the drive down. Probably why a number of modern flow hoods and benches use electronic methods over mechanical baffling, despite it being more time consuming and costly to design, pass through CE testing, produce and fit than a sliding bit of metal.

A 'vented soil pipe' 'access pipe' (£8.94 from toolstation.com in the UK, free next day delivery on orders of £10, free hot drinks in the store & the lowest prices available, sweeet);


There are also other handy bits available. Such as prebent angle bends (which are helpfully made with a slight off 90 degree angle to encourage draining) or this, an adjustable bend to take up errors;


Flexible couplings;


These ones are cheaper and come in lots of different configurations to couple from 110mm to various other sizes. They produce the same water / air tight seal and have big hoze clips to get them tight, as you can see (use a ring spanner or socket set for ease, screwdrivers are a pain in the ass).


There a many more, and the way you set the up is largely limited by your inventiveness.

What you've linked to is essentially a wash head. If you've seen someone going on about them on this forum, there's a good chance it was in one of my posts, as I rant on about their virtues all the time.

The only difference between mine and the one you've shown is that mine is all glass, because they're also used for analytical, experimental work that may interact with the rubbers and such in that one. E.g. HCl(g) will slowly rot orange natural rubber and DCM will make it swell up. But you can get inventive with that two and line the stoppers with PTFE tape or use silicone or poly stoppers.

If you're just using it for scrubbing, I expect that one will do you fine. Along with a number of variations on it that people have made with jars. Fritted glass is better, but the glass frits also have a tendancy to fall apart in strong base solutions, and the aquarium ones can melt in some solvents or corrode around things like HCl(g).

I would hazard a guess that the majority of nasty things you'll produce will be either going acidic or basic when in solution and likely soluble in water; HCl(g) is very soluble, for instance. To scrub HCl(g), you could take that bottle you linked to and simply fill it with extremely concentrated KOH. With no frits to dissolve, it'll sit there happy as piggies in shit (which aren't actually that happy in actual shit) until you're too old to do chemistry anymore. Any HCl(g) that attempts to get through it will immediately be raped, hard, to potassium salt. Use NaOH and you've got table salt. There will essentially zero HCl(g) at the output.

So, selecting to have a frit or not and the material it's composed of depends on the wash in progress. In some cases, plastic will work better than glass. In others, the plastic will be a molten blob on contact.

You can also stack a number in series to scrub a lot of different possibilities, or increase the efficiency again.

I'll include a picture of the wash heads & bottles QuickFit here in the UK do at the end. This one is a sintered version.

All thar best!
John

{edit}I just noticed the price on that bottle you linked to. YOWZERS! That's a good 10 to 15 times cheaper than the bottle alone for this one, used.



[Edited on 6-8-2010 by peach]

bob800 - 5-8-2010 at 17:35

First of all, thanks for writing all that info!

So you're saying I could use something like this: http://www.heavenfresh.com/Replacement-hepa-filter-for-XJ-38.... Or would something like this be better: http://cgi.ebay.com/4-x-12-ACTIVATED-CARBON-CHARCOAL-FILTER-...

bob800 - 5-8-2010 at 17:40

Quote: Originally posted by peach  

{edit}I just noticed the price on that bottle you linked to. YOWZERS! That's a good 10 to 15 times cheaper than the bottle alone for this one, used.


Yeah, I know elemental scientific is REALLY cheap. They have 500 ml ACS Nitric acid for $9.99!!! Only problem is that they're REALLY slow at filing orders.

peach - 5-8-2010 at 17:48

The cylinder will likely be easier to fit into the system.

The HEPA is likely to have finer powdered material in it, whereas that cylinder has granules (I think mine had powder, not sure).

As to which is better for your specific design, that's really something you'll need to have an investigate into, such as checking the recommended fans, CFMs and pressures of those or other alternatives and thinking about how your want to lay it out or vent it. You can make rough comparisons by looking at references like 3M, who make a lot of the HEPAs and list their flow characteristics. Filters of a similar size from different companies will have roughly similar figures attached. The same applies for the cylindrical filter, there will be people selling them who list the flow rates at different pressures. Asking the stoners at the cannabis growing forums would be a good place to start, as soon as they wake up and stop watching daytime TV, eating packs of biscuits.

The micronic filtering nature of the HEPAs makes me suspect the flow rate of that granule only cylinder will be higher if they have roughly the same surface area. You can work that out from the diameter and compare it to the flat sheet filter.

[Edited on 6-8-2010 by peach]

peach - 6-8-2010 at 04:39

I'd still consider the PVC pipe option. With the pipe, you'll know what all the seals and such are so you'll better know their corrosion properties. PVC is pretty good in terms of that. The mild steels will start corroding if things like HCl(g) go through. Some of the others components may as well. I suspect granules will do if you fill most of the access pipe. If you can still smell anything, add a handful of PAC.

Unless you've dealt with things like HCl(g), you'll probably underestimate it's ability to rot things. As an example, I can have all my tapers either greased or sealed with PTFE tape and keck clipped. If a few tiny puffs of it escape whilst I'm swapping something around, or if the grease isn't well seated or runs out, things five to ten metres away will start going. My cooker hood is about that far away from the surface I produced some HCl(g) on a few weeks ago. Despite the amounts of gas escaping being so low they are barely detectable, there was rust appearing on the stainless a week or two later. Similarly, if I clean the glass out over the kitchen sink, there will inevitably be a few traces of something left and it'll fume a little. Not enough to be dangerous, but the fumes have rotted a brand new stainless sink to look like it's a decade old, despite me blowing them away and thoroughly rinsing down the surfaces. In fact! I just remembered I have something I can show you...

This is a pair of stainless steel scissors I had lying on the bench beside me when a small, virtually undetectable amount of HCl(g) escaped the glass, they were shiny prior to this;


Note how the opposite side is significantly less corroded, indicating just how sensitive things like this are to the flow rates and directions (it's only stained the top, where it could land, like the germs in Pasteur's broth flasks [Pasteur actually 'stole' a lot of the limelight for germ theory from the earlier Semmelwies, who was told he was nuts for thinking tiny animalcules were everywhere and he died at a young age in a psychiatric hospital, after publishing reams of papers regarding their probable existence, his evidence based control strategies and the beneficial results this achieved in hospitals]);


Despite them having stainless stamped on them, that term is strongly bastardized now to mean anything with any chrome, moly etc content in it that exceeds mild steel grades, so a lot of the 'stainless' you find around the house isn't actually all that stainless. You need graded, controlled stainless like 316, Inconel and the others. But it's a fair indicator of how aggressively things like this will attack metals.

In a lot of cases with chemistry, building your own is usually going to end up costing more in money and time, performing unacceptably or being a risk. I think the access pipe scrubber is reasonable unique in that it is likely to be cheaper, quicker, work just as well or better and present well known and controllable variables.

John

[Edited on 6-8-2010 by peach]

bob800 - 8-8-2010 at 08:37

Do you think my blower will corrode too then?

peach - 8-8-2010 at 08:44

Quote: Originally posted by bob800  
Do you think my blower will corrode too then?


If the blades are mild steel and HCl(g) comes into contact with them, yes.

There's a good chance they'll look like those scissors after it touches them the first time.

But with iron, the corrosion can actually help retard further corrosion provided the layer isn't being mechanically wiped or washed off. Out in the wild, all the rain and things moving around will tend to rub the layer off, which then reforms, gets rubbed off, et cetera until the surface is eaten through. In your fan, hopefully, the air flow alone won't be enough to do any significant rubbing. You may open it and discover the blades are brown, but still perfectly intact.

People who paint things commercially purposefully spray all their iron and things with that universal red paint prior to the colour layers. That's not only a primer / mask, it's red iron oxide and there as a corrosion inhibitor; the poor man's galvanizing.

It's also a bit like cutting metal with a flame torch, plasma cutter or laser. The heat won't really cut through it alone, the molten blob needs blowing through with gas; mechanical agitation.

Same is true etching PCB's, they want bubble tanks or pumps in the tank to mechanically replace the diffusion gradients and rub the metal coming off away.

Something with PVC blades would probably come out looking cleaner, but as I say above, a bit of surface muck isn't so bad.

The bearings will almost certainly be stainless steel, and probably closed, so they shouldn't be too much an issue.

[EDIT]I thought of a cheap alternative, or one that'll save you the effort of looking for another fan and doing all the curves again anyway, just paint the blades with something like enamel. If you're bothered about the appearance of things and the seals, again, you'll want to use 110mm or bigger soil pipe, the mild steel will rot (not through, but it'll look shitty). Any slight fails in the seams may get the steel going brown on the outside as well; it only takes a tiny, tiny amount. I can't imagine you'll get a comparable seal on the steel without siliconing it all. Which can also be messy and come apart with movement. Those 110mm seals take seconds and are easily tight enough for gas at such low pressures (I needed big hydraulics to pull one open). You can even wrap a layer of PTFE plumbing tape over the OD of the male pipe and it's ends first for a pretty much chemically invulnerable seal.

[Edited on 8-8-2010 by peach]

bob800 - 8-8-2010 at 09:01

.

On the grainger page, (http://www.grainger.com/Grainger/wwg/search.shtml?searchQuer...) it says the "Housing Material" is rolled steel and the "Housing Finish" is gray enamel. But I don't think that's referring to the blades. Can you tell what they're made of from the picture?

The enamel paint idea sounds good, but I don't want to paint the blades shut!

peach - 8-8-2010 at 09:11

That's excellent that it's already painted with enamel, you're saving some spray painting there.

I would be very impressed if you managed to spray the blades shut, it'll be next to impossible. I guess you may have been thinking of using a brush perhaps?

Nah, don't bother. They sell enamel in spray cans at the hardware and car places. It's often called 'radiator paint'.

Take the cover off the fan and try to get the blades and the motor separate from the housing. It'll be super, duper easy to paint something like that, even easier than most things. Just give the fan a spin (with a stick, not the socket) and spray as it turns. You may want to undo the set screw and take the blades off entirely to do the back as well. Remember to keep the can 12" or so away, give it a good 2 minute shake first, put on a damp coat and add more, carefully as per the instructions. If you put your greasy mits all over the blades and don't follow the time intervals on the pack, it will mess with the paint. Remember there are 'hidden spots' as well, you need to be ducking and leaning all over it under a bright light to find where you've missed. Spraying outside usually wastes a lot of paint.

I can't really tell what the metal is from that picture, it's bright, but it's also brand new. Stainless is quite a bit more expensive to buy and process, so it's more likely it's mild steel. They may have galvanized it, but I'd have to have it in my hands to see that ideally. If it's zinc galvanized, a number of paints won't stick to that without a special primer, so get the fan first or hold onto your spray can receipt.

[off topic]I used to work in a metal fabricators. For one job, they hot dipped a bunch of supports for a railing fence at an airport. Then bought the special acid etch primer. Then were told by the spray guys one drum of the expensive commercial stuff they'd bought would do the entire job with most of it left over. They had about ten on the van. Ask first, buy later. :P

E.g. Email them and ask if someone can try sticking a magnet to the blades.

[Edited on 8-8-2010 by peach]

bob800 - 8-8-2010 at 19:07

OK, I'll go with the epoxy idea. Thanks for all your suggestions.

Attached is my "final" plans for the hood, along with the dimensions.

Also, the depth doesn't have anything to do with how many cfms I need, right? The only thing that matters is the face velocity, right? How deep do you think I should make it?

[slightly off topic]. You said you're using a 1L RBF, right? I don't think I need that much capacity though. Do you think this 19/22 setup would fit in my 2.5' sq hood: [url=http://cgi.ebay.com/New-19-22-Chemistry-Laboratory-Distilling-Kit-FREE-S-H-/200383943175?pt=LH_DefaultDomain_0#ht_3129wt_1137]

Again, thanks for all your help.

plans.bmp - 748kB

anotheronebitesthedust - 9-8-2010 at 01:00

Do they make blowers with ptfe or other non-corrosive, highly abuse resistant materials without charging $1000's?

I found this site but they are located in China and I have a hard time trusting asians. And I still wasn't able to find a worthy model.
http://www.blower.com.tw/en_pdc00.htm

anotheronebitesthedust - 9-8-2010 at 01:50

Well found what I was looking for from Grainger, but as I thouth the price would be in the $1000's.


http://www.grainger.com/Grainger/PLASTEC-Blower-3MPW3?Pid=se...
But one benefit is that the noise level is only 43 dB

peach - 9-8-2010 at 17:59

Quote: Originally posted by anotheronebitesthedust  
Do they make blowers with ptfe or other non-corrosive, highly abuse resistant materials without charging $1000's?

I found this site but they are located in China and I have a hard time trusting asians. And I still wasn't able to find a worthy model.
http://www.blower.com.tw/en_pdc00.htm


You can forget that matey! :P

PTFE = big cash even for objects that can be easily made from standard solids without excess machining or special moulding methods. A centrifugal fan blade is reasonably complex by comparison and theres basically no demand outside of very odd chemical handling problems, meaning they can charge whatever they like, leaving you with you're anus stinging.

I think it might be worth modifying your attitude towards Asians a little there. I get your point, that it's easy to get ripped off from those distances and be able to do nothing about it. However, I have ordered quite a few items from China and actually had the guy send it quicker, cheaper, better and with less hassle than people in my highly developed home town of England.

Also remembering, China is a highly mechanized, technology leading country (a ton of the big name US / European countries have plants out there, not at home). If you pay for rubbish, they'll give you rubbish. If you pay a decent amount, they can produce equally good products, or better (given that 'they' work a lot harder most of the time). As you can see from the photo, I am not 'one of them'. ;)

Besides, if you scrub the gas stream a little prior to it hitting the fan, as per Bob's suggestion in his 3 year old looking paint diagram :D (jk), most of the corrosives / solvents should be massively diluted down before they get to the blades.

@bob800

>>>>A video reply!<<<<

[Edited on 10-8-2010 by peach]

bob800 - 9-8-2010 at 18:58

Wow, thank you so much for making that!:). I guess it makes up for your MS paint insult!:D

[EDIT] I just bought the wood, plexiglass, and blower for the hood. It's going to be 2.5' sq with 2' depth. Hopefully that'll fit a 19/22 setup...

[Edited on 10-8-2010 by bob800]

peach - 9-8-2010 at 19:43

Glad you liked it. Pictures are a lot more useful most of the time. I'll be stalking you for photos of said fume hood. ;)

[Edited on 10-8-2010 by peach]

cnidocyte - 11-8-2010 at 03:00

I nearly have my fume hood built, just the hard part left. The fan. One of the gypsum walls of my hood faces the garden so I could literally just install a box fan right into the wall of the hood but I'll have to cover the motor to make sure it doesn't ignite any flammable fumes. I'm definitely gonna use solvents like ether and acetone in my hood. I have ADHD and even on dexedrine the main fume hood thread is painful to read so I don't think theres any harm in having more than one DIY fume hood thread going.

I built the floor of my hood with these old boards I had lying around, I can't remember what the name of this material is but whatever it is it literally as no chemical resistance, I tested a spare bit of board with some NaOH solution and a day later the insides of it were oozing out. I need to cover this crap in something chemical resistant. I put down some polypropylene mats (they sell them for putting in drawers).

[Edited on 11-8-2010 by cnidocyte]

peach - 11-8-2010 at 05:23

I have to bring out my negative nancy hat here and say, that's a bit of a poor excuse. If you can't read the thread, it kind of suggests you might not be too careful using things that require the hood in the first place.

Acetone isn't much of a fire problem. I smoke and play with my lighter sat right in front of it, with it all over the glass (as I rinse it), the surface, the paper I'm writing on, tissues, rags, me... I've never seen it ignite, ever, and I've gone through tens of litres of it.

Ether is another issue altogether.

That's where I'd start considering the benefits of a sparkless motor.

With the doors and windows open as I work with misbehaving solvents, the escaping vapour is rapidly diluted by air passing through. In a hood, it'll puddle up and all have to go out past the fan.

The only time I've had a fire was with cyclohexane. Again, I'd been sat smoking with it all over the place all night. Then someone else came into the room, tried to light a candle and the piece of paper a few drops had landed on immediately burst into flames; within 30s of them coming into the room. Indicating that it's not so much the ignition source, as who's controlling it.

I'm clinically insane (quite literally, according to the NHS). They've had me on their special person pills, checking in for hour long chats with psychiatric teams every two weeks for a year and I only mention it because you mentioned ADHD. I can still read the threads and think they should be merged. :P

[Edited on 11-8-2010 by peach]

entropy51 - 11-8-2010 at 05:39

Quote: Originally posted by peach  

Acetone isn't much of a fire problem. I smoke and play with my lighter sat right in front of it, with it all over the glass (as I rinse it), the surface, the paper I'm writing on, tissues, rags, me... I've never seen it ignite, ever, and I've gone through tens of litres of it. .....

The only time I've had a fire was with cyclohexane. Again, I'd been sat smoking with it all over the place all night. Then someone else came into the room, tried to light a candle and the piece of paper a few drops had landed on immediately burst into flames; within 30s of them coming into the room. Indicating that it's not so much the ignition source, as who's controlling it. ........

I'm clinically insane (quite literally, according to the NHS).

peach - 11-8-2010 at 05:49

L@@K @T>>>>>Yet again, he pounces on the chance to have a complain and ...'s out the reasoning to shape it to how he likes it. This seems to be a very common, annoying, theme for yourself. You'll actively seek out things to have a complain about when people try to make the offer bit of help (I've seen you doing it in other people's thread as well, over very minute details), you'll tell me how to word my replies, then you'll actively edit them to fit your own opinion. The people around you in the real world must be extremely patient. Sometimes, you need someone else to point these things out to you. Try making your own videos, and you'll see how you appear to the rest of us.<<<<<<I makes you a special video reply, 'dude'!

[Edited on 11-8-2010 by peach]

cnidocyte - 11-8-2010 at 06:21

I read the thread, I just said it was painful. I'm clinically autistic according to the psychiatrists but you can't believe everything they tell you. I'm a bit less meticulous with the fan part of my hood because I could literally take a wall off it and use 1 of those portable oscillating fans to get rid of the fumes.

peach - 11-8-2010 at 06:33

Unless you're speaking to a consult, who actually has a degree, you're probably talking to an idiot who doesn't know anything about what pharmaceuticals are, how they work, what neurotransmitters are and who probably can't name a single region of the brain. I can describe it in detail, and I'd still feel uneasy telling others what they're thinking or how it's working.

Ether is a fair bit worse than acetone, so you need to be careful using ultra cheap motors; as they tend to emit visible sparks continuously as they spin. Not including a scrubber before the fan will make things worse (as I say in my reply to bob, the fumes will be going straight through the motor). As a lot of those motors use the air going past to cool themselves, that's not great at all when something's potentially flammable.

Centrifugals at least have some degree of isolation between the motor and the blades. An oscillating fan, or wall fan, will often suck vapours back in as they go past. If you can see 'grids' on the sides of the motor body, don't use it. Vacuum cleaner motors do this as well. Although, if you can put up with listening to one of those for 6 hours, 3rd degree burns will likely be a godsend.

Like... oh my godz... fire!!!!!!

Ideally, everyone should have a CO2 & powder extinguisher, as they'll do just about all fires when combined. But particularly so if you're not confident around flammables. You need the big kind as well. The small one's will barely cope with a B14 spill. I can promise you, my tiny acetone fire is nothing like the reality of it all going up unexpectedly, you will want an extinguisher and the adrenaline will be rushing harder than you've felt it for most of your life. And it will mean the difference between you loosing your home, people being scared for life or dyling and just getting a fright. Best way to start understanding what you're dealing with, try setting fire to it on purpose, as I did above. Then you'll appreciate, tangibly, how quickly it boils off and how it ignites.

[Edited on 11-8-2010 by peach]

Neil - 10-5-2011 at 16:45

The thread was kinda wandering towards the end... but my 0.02$


Poly-Carb plastics are very poor performers around some esters and solvents.

Case and point, take a used a glowstick and toss it in a Poly-Carb bottle, then seal the bottle. In a week or two the bottle should shatter like glass if dropped from any height.

I've also seen solutions of dish soap crystallize name brand Poly Carb so bad that it began to crack and splinter apart. Plain Plexi glass under the same conditions remained unchanged.

In building a hood it may serve you better to go with cheaper plastics then the expensive Poly Carb.


My two cents..

albqbrian - 30-5-2011 at 02:41

I'll definitely be building one when I finally return to the US, and my house.

The room I'd set it up in has swamp cooler provided cooling. Maybe I could figure a way to reverse that???

One thought on PVC. As I recall from my woodworking days, moving stuff through PVC can build up some serious static electricity. You're always warned to ground such systems. I don't know if that would be an issue in the ones discussed here though.

Do most of you just run it out a convenient window? I do have one thing in my favor. I'm just downhill from an extremely large, hitech manufacturing plant. One that some folks in my town are always berating/accusing/investigating this company for releasing all sorts of bad, smelly chems. At least I won't get blamed :P

watson.fawkes - 30-5-2011 at 03:25

Quote: Originally posted by albqbrian  
One thought on PVC. As I recall from my woodworking days, moving stuff through PVC can build up some serious static electricity. You're always warned to ground such systems. I don't know if that would be an issue in the ones discussed here though.
The bulk of static buildup in woodworking system arises because you're moving a particle-laden stream through the pipe. Particle-to-wall friction transfers charges; the air stream then moves charged dust. If you're not moving dust, charge buildup is very much less a problem.