Sciencemadness Discussion Board
Not logged in [Login ]
Go To Bottom

Printable Version  
Author: Subject: Homemade spectrometer
Quibbler
Hazard to Self
**




Posts: 65
Registered: 15-7-2005
Location: Trinidad and Tobago
Member Is Offline

Mood: Deflagrated

[*] posted on 3-8-2016 at 03:15
Homemade spectrometer


Many years ago I was involved in a thread here to measure the absorbance of visible light for a Beer-Lambert type concentration determinations.
I had some success using LEDs (with tight wavelenghts).

Now I have made a spectrophotometer that will go form 400 to 900 nm and has a resolution of about 4 nm.

I have attached a picture it shows a test spectrum of a didymium filter (used for wavelength calibration). If you would like to check the spectrum here is a link to the expected spectrum.
didymium glass reference

For scale you can see a 1 cm cuvette on the right below the screen and sticking out of the left sde is a micro SD card (should I need to store the spectrum).



P1030132.jpg - 84kB

[Edited on 3-8-2016 by Quibbler]
View user's profile View All Posts By User
Marvin
National Hazard
****




Posts: 995
Registered: 13-10-2002
Member Is Offline

Mood: No Mood

[*] posted on 3-8-2016 at 04:30


That looks impressive. Can you post build details?
View user's profile View All Posts By User
Quibbler
Hazard to Self
**




Posts: 65
Registered: 15-7-2005
Location: Trinidad and Tobago
Member Is Offline

Mood: Deflagrated

[*] posted on 3-8-2016 at 05:27


Attached is a photo with the top removed showing the works.
The main items are:
Xenon torch bulb
Diode array (1x256) TSL1402 (from Digikey)
PIC16F1789 microprocessor (large RAM needed)
The diffraction grating is one of those plastic holographic ones (1000 lines) which are cheap.
The lenses (two) are plastic 3cm focal length.
The display is colour TFT 320x240 which are very cheap (see S.K.Pang for example) some have an integral SD socket (mine didn't).
Actually a B/W display would have been better, but you can't get them (the colour display is slow and sucks power).

P1030133.jpg - 454kB
View user's profile View All Posts By User
Marvin
National Hazard
****




Posts: 995
Registered: 13-10-2002
Member Is Offline

Mood: No Mood

[*] posted on 4-8-2016 at 00:47


That's a really cool build and kudos for using veroboard. That's a proper old school prototyping/one off method. What's the resolution of your ADC? have you tried doing Beer Lambert stuff?
View user's profile View All Posts By User
Quibbler
Hazard to Self
**




Posts: 65
Registered: 15-7-2005
Location: Trinidad and Tobago
Member Is Offline

Mood: Deflagrated

[*] posted on 4-8-2016 at 02:58


The 16F1789 has a 12-bit ADC. As the collection is very fast I normally collect 256 scans then throw away the least significant 4 bits and store the reading as a 16-bit number. This is converted into absorbance using a lookup table.

One of the problems of using a diode array is stray scattering from the diffraction grating can cause problems. There is a routine that tries to compensate for this, but it doesn't always achieve perfection.
The B-L plots are good up to Abs of 1.5 then a longer intergration time is needed on the diode array.

Attached is the spectrum of a real sample potassium permanganate. This is one of the few compounds that have a marginally interesting visible spectrum.

PMAN256B.bmp - 120kB

[Edited on 4-8-2016 by Quibbler]
View user's profile View All Posts By User
Texium
Administrator
Thread Moved
27-11-2023 at 10:45

  Go To Top