CONSTRUCTION OF WATER ASPIRATOR PUMP
Materials required:
Harbor Freight, Item 2955-7VGA 3/4 HP, 1" Cast iron shallow well pump, $129.99, Pacific Hydrostar, 20-60 psi
OR
Item 47906-1VGA, 1HP, 1" Shallow well pump with stainless steel housing, $99.99, 72 psi
Stainless steel is best option in my humble opinion,
Water tank considerations, use 55 Gallon plastic trash container with lid, plastic is chosen due to corrosive pump
gases, Et. cetera, plastic lid chosen due to ease of inlet and outlet hole construction, and ease of ice addition, control of dirt, leaves....
Water Aspirator Pump, plastic or metal
Wards Natural Science Co. $17.80
Cole-Parmer (plastic) $15.25
Avogadro's Lab Supply (ebay).......$24.99
Construction of apparatus
use one inch intake and output piping, use stainless or plastic, use valves on both intake and especially output, place output valve after tee where
aspirator is located, thus water pressure can be regulated through aspirator, do not place pump on plastic lid.....
Place holes in plastic lid, three are required, one for intake, one for aspirator, and finally one for bypass on the pressure side. Use large linear
hose to control output of aspirator or gross splashing will occur, construct aspirator pump device in open area, i.e., outside.... and protect from
freezing.
I will post answers to questions concerning this construction...
|
|
bbartlog
International Hazard
Posts: 1139
Registered: 27-8-2009
Location: Unmoored in time
Member Is Offline
Mood: No Mood
|
|
While this works, isn't it quite inefficient compared to buying a vacuum pump?
Also, why a 55 gallon water tank rather than just a five gallon one?
|
|
watson.fawkes
International Hazard
Posts: 2793
Registered: 16-8-2008
Member Is Offline
Mood: No Mood
|
|
This plan is for a
high-volume pump, of the kind used on a scale they call "pilot plant", the kind you use with a 5L reaction flask. While the ultimate pressure is no
better than any other aspirator pump, it will pump higher volumes of gas.
|
|
gsd
National Hazard
Posts: 847
Registered: 18-8-2005
Member Is Offline
Mood: No Mood
|
|
Because the higher the volume of recirculated water, lower the rise in its temperature and hence lower the drop in vacuum.
gsd
|
|
aonomus
Hazard to Others
Posts: 361
Registered: 18-10-2009
Location: Toronto, Canada
Member Is Offline
Mood: Refluxing
|
|
Quote: Originally posted by gsd |
Because the higher the volume of recirculated water, lower the rise in its temperature and hence lower the drop in vacuum.
gsd |
Also probably the less chance of bubbles from the aspirator getting sucked into the pump, they will have a chance to float back to the surface due to
a lower flow velocity in the bulk of the fluid with a larger container radius.
|
|
hissingnoise
International Hazard
Posts: 3940
Registered: 26-12-2002
Member Is Offline
Mood: Pulverulescent!
|
|
Drug chemists have been using this kind of set-up for years - they work well as long as outlet bubbles are kept out of the pump-inlet.
And the bigger the reservoir the better!
A small amount of ordinary antifreeze will protect against frost and the corrosion inhibitors will afford some protection for the pump. . .
|
|
poppy straw
Harmless
Posts: 4
Registered: 20-8-2009
Member Is Offline
Mood: OLD
|
|
EXCELLENT QUESTIONS
The water aspirator pump as configured with LARGE volume of water resolves two problems, the first is function of use versus increase in water
temperature, maximum vacuum is achieved with minimum water temperature, if for example using a rotary flash evaporator will increase water temperature
of the pumping system, addition of ice will help... acid neutralization by hydroxide addition to pump water... antifreeze addition.... complete
control of contaminated pumping water...,
Additionally, toxic and corrosive gasses are pulled through the system and are dissolved or released to ambient environment, this is why the pumping
aspirator system is used at all university laboratories.... for general lab. functions like filtration.... distillation, etc.
Expensive oil driven vacuum pumps require cold traps to freeze out harmful vapors from the vacuum stream...corrosion is very harmful to mechanical
vacuum pumps.. cold traps requirements are expensive, dry ice/acetone, alcohol, liquid nitrogen, and dewar vessels, etc.
Then, for practical purposes, the following ranges of pressures are distinguished:
water pump vacuum (ca. 10-700 mm)
oil pump vacuum (0.001-1 mm)
mercury pump vacuum(<10 to neg. third mm)
see, Organicum, B.J. Hazzard pages 25-32 for excellent discussion. isbn 0-201-05504-X or there a bouts
|
|
anotheronebitesthedust
Hazard to Others
Posts: 189
Registered: 24-6-2007
Member Is Offline
Mood: No Mood
|
|
Any pictures?
|
|
gsd
National Hazard
Posts: 847
Registered: 18-8-2005
Member Is Offline
Mood: No Mood
|
|
Quote: Originally posted by aonomus |
Also probably the less chance of bubbles from the aspirator getting sucked into the pump, they will have a chance to float back to the surface due to
a lower flow velocity in the bulk of the fluid with a larger container radius. |
Normally the bubble menace is tackled by proper partitioning / baffling of the receiver tank so that the suction to the pump and the input from
ejector are in two separate sections.
gsd
|
|
Panache
International Hazard
Posts: 1290
Registered: 18-10-2007
Member Is Offline
Mood: Instead of being my deliverance, she had a resemblance to a Kat named Frankenstein
|
|
Just use mains water straight down the drain, no cooling worries, no pump required, might want a cold trap to guard against accidents, then sit back
in your baby seal fur jacket, turn on your CFC driven air-conditioning on full to compensate for having no insulation and watch the megalitres piss
down the drain. Quite a show.
|
|
Magpie
lab constructor
Posts: 5939
Registered: 1-11-2003
Location: USA
Member Is Offline
Mood: Chemistry: the subtle science.
|
|
Quote: Originally posted by Panache | Just use mains water straight down the drain, no cooling worries, no pump required, might want a cold trap to guard against accidents, then sit back
in your baby seal fur jacket, turn on your CFC driven air-conditioning on full to compensate for having no insulation and watch the megalitres piss
down the drain. Quite a show. |
Or, if in the northern hemisphere, substitute "wall mounted electric heater on full" for "CFC driven air-conditioning on full."
Your mention of the cold trap is a good point, Panache. Just the other day I was attempting to remove some ethanol from a diethyl malonate
esterification by vacuum distillation at 50C when, being a bit casual with vacuum management, I had a massive suckback. I normally don't use a safety
bottle and this poor practice finally caught up with me. So I consulted Vogel, and here is what I came up with as a safety bottle. It works very
well and by using the Hoffman screw clamp on the air bleed I can fine tune the pot vacuum to within a gnat's eyelash:
The single most important condition for a successful synthesis is good mixing - Nicodem
|
|
poppy straw
Harmless
Posts: 4
Registered: 20-8-2009
Member Is Offline
Mood: OLD
|
|
These images might be helpful:
|
|
Panache
International Hazard
Posts: 1290
Registered: 18-10-2007
Member Is Offline
Mood: Instead of being my deliverance, she had a resemblance to a Kat named Frankenstein
|
|
Quote: Originally posted by Magpie | Quote: Originally posted by Panache | Just use mains water straight down the drain, no cooling worries, no pump required, might want a cold trap to guard against accidents, then sit back
in your baby seal fur jacket, turn on your CFC driven air-conditioning on full to compensate for having no insulation and watch the megalitres piss
down the drain. Quite a show. |
Or, if in the northern hemisphere, substitute "wall mounted electric heater on full" for "CFC driven air-conditioning on full."
Your mention of the cold trap is a good point, Panache. Just the other day I was attempting to remove some ethanol from a diethyl malonate
esterification by vacuum distillation at 50C when, being a bit casual with vacuum management, I had a massive suckback. I normally don't use a safety
bottle and this poor practice finally caught up with me. So I consulted Vogel, and here is what I came up with as a safety bottle. It works very
well and by using the Hoffman screw clamp on the air bleed I can fine tune the pot vacuum to within a gnat's eyelash: |
How do you bore out your rubber stoppers? No method i have tried seems universally foolproof. Do you fix the stopper in a sleeve or have it just
clamped? Hole saw, spade bit, drill bit. I finally figured out a great way to seal a teflon thermocouple wire through a rubber stopper, drive a thick
syringe needle through, feed the wire through the needle, remove needle, voila, not at leak to be seen.
I always try not using anything other than glass or polyolefins in the safety trap/cold trap, that way i never have to consider its survival from exp
to exp.
I took a photo quickly but the labs so messy i was embarrassed, i'll post a photo tom, but in short my setup is very adhoc but works great.
Is that galvanised iron against the wall, i'm too sloppy (my minds on more important things) to have anything other than tiles, my sink always ends up looking like an old tinnie before too long.
[Edited on 19-1-2010 by Panache]
|
|
aonomus
Hazard to Others
Posts: 361
Registered: 18-10-2009
Location: Toronto, Canada
Member Is Offline
Mood: Refluxing
|
|
I have found that sticking the stopper into a freezer for a few days, then quickly taking it and drilling it does an ok job. I have a rubber stopper
sized for a 24/40 joint, with a K type thermocouple probe running through it (metal insertion probe). No leaks with hot solvent yet surprisingly...
|
|
Magpie
lab constructor
Posts: 5939
Registered: 1-11-2003
Location: USA
Member Is Offline
Mood: Chemistry: the subtle science.
|
|
Quote: Originally posted by Panache |
How do you bore out your rubber stoppers? No method i have tried seems universally foolproof.
I always try not using anything other than glass or polyolefins in the safety trap/cold trap, that way i never have to consider its survival from exp
to exp.
Is that galvanised iron against the wall, i'm too sloppy (my minds on more important things) to have anything other than tiles, my sink always ends up looking like an old tinnie before too long.
[Edited on 19-1-2010 by Panache] |
>I use a regular cork boring device, with the different size cutters for different size holes. It works only so-so for rubber stoppers. Use of a
drill on a frozen stopper sounds interesting - I'll have to try that next time.
>My use of copper is unorthodox but not without reasons: (1) The only stopper I had that size already had a large hole in it. A 1/8" brass pipe
nipple fit it perfectly giving a good seal. (2) I wanted to use the rugged brass hose barbs at the connections as this flask must also server for
filtrations and therefore must be disconnected/reconnected occaisionally. (3) I needed this right away and it was all I could cobble together the
same day.
>That wall is not galvanized iron but is the outside of my aluminum fume hood. It looks like my fabricator used an orbital sander on it. I don't
know why he did this. My sink and aspirator are about 3 meters away.
[Edited on 19-1-2010 by Magpie]
[Edited on 19-1-2010 by Magpie]
The single most important condition for a successful synthesis is good mixing - Nicodem
|
|