Author: Subject: Data Validation of previous Eosin Y results for April/21/2016
dhaffnersr
Harmless
*




Posts: 30
Registered: 29-3-2016
Location: Hopewell VA
Member Is Offline

Mood: always cool and calm...

[*] posted on 23-4-2016 at 11:18
Data Validation of previous Eosin Y results for April/21/2016


Using my Aries series green laser, I wanted to validate the Eosin Y study I did the other day with the Prolight Uv LED (410nm). I'm presenting this study as a final result for this series of tests using my plab v2.5 spectrometer (build 3.0)

Sample preparation:
2 cuvettes, one for the Blank (0.1M NaOH in 20 % ethanol) and one for samples S1 thru S6.
Volume capacity calculations- 1 drop of eosin y (1 % aqueous solution) = 0.058g
To convert to micro liters take 0.058g*1000 = 58 microliters
Sample(1)-1 drop 58 microliters
Sample(2)-2 drops 127 microliters
Sample(3)- 3 drops 189 microliters
Sample(4)- 4 drops 254 microliters
Sample(5)- 5 drops 320 microliters
Sample(6)- 6 drops 372 microliters
Each scan will last 30 seconds.
Excitation source- 532nm Aries series green laser (20mW)



green laser3.png - 829kB PLOT 1 eosin Y apr22.png - 187kB PLOT 2 eosin Y apr22.png - 256kB emmssn eosin y.png - 27kB 2rawdata eosiny S1 apr22.png - 41kB S1 thru S6 532nm eosin y apr22.png - 34kB S1 thru S6 with BLANK plot apr22.png - 32kB
View user's profile Visit user's homepage View All Posts By User
dhaffnersr
Harmless
*




Posts: 30
Registered: 29-3-2016
Location: Hopewell VA
Member Is Offline

Mood: always cool and calm...

[*] posted on 23-4-2016 at 13:52


More on my research can be found here:

http://publiclab.org/profile/dhaffnersr

;)

View user's profile Visit user's homepage View All Posts By User
nezza
Hazard to Others
***




Posts: 324
Registered: 17-4-2011
Location: UK
Member Is Offline

Mood: phosphorescent

[*] posted on 24-4-2016 at 23:47


Interesting work, and some of your publiclab studies look good as well. I have recently been doing some spectrometry (Not as comprehensively as yourself) on some fluorescent dyes including Eosin Y. The spectrometer I use is a science-surplus unit using spectrum studio software. Using a white light source to generate an absorption spectrum I find a spectrum shift on fluorescent materials with concentration. I am assuming this is due to some self fluorescence from the dye caused by the white light source. I know Eosin's maximum absorption is in the green so I will try to obtain an emission spectrum of the dye. I have attached a plot of the curve at different concentrations. I suspect a scanning spectrophotometer using monochromatic light would be the best way to obtain a true absorption spectrum.

Eosin.jpg - 144kB




If you're not part of the solution, you're part of the precipitate.
View user's profile View All Posts By User
dhaffnersr
Harmless
*




Posts: 30
Registered: 29-3-2016
Location: Hopewell VA
Member Is Offline

Mood: always cool and calm...

[*] posted on 25-4-2016 at 02:34


Hey nezza, nice plot. Remember that at higher sample concentrations your going to start suffering the "inner-filter" effect, I usually start with 6 different sample concentrations, this usually gives the best calibration curves.

Yes, you are correct about using the monochromatic light, I use a UV laser (405nm) and a green laser (532nm) for my excitation light sources.:cool:
View user's profile Visit user's homepage View All Posts By User
nezza
Hazard to Others
***




Posts: 324
Registered: 17-4-2011
Location: UK
Member Is Offline

Mood: phosphorescent

[*] posted on 26-4-2016 at 23:21


Thanks for the info. I have attached my traces of 3 dyes using a 405nm laser for the fluorescein and a 532nm laser for the other 2. Th traces are much cleaner.

Dyes.jpg - 169kB




If you're not part of the solution, you're part of the precipitate.
View user's profile View All Posts By User
dhaffnersr
Harmless
*




Posts: 30
Registered: 29-3-2016
Location: Hopewell VA
Member Is Offline

Mood: always cool and calm...

[*] posted on 27-4-2016 at 03:36


hey nezza:cool:

very clean, what are the peak wavelengths? curious because I have to use Plab's spectral workbench 2.0 software online to run my hardware to capture spectra and they run java script, and I think there are inconsistencies in my results due to the way they average the pixels.
View user's profile Visit user's homepage View All Posts By User
nezza
Hazard to Others
***




Posts: 324
Registered: 17-4-2011
Location: UK
Member Is Offline

Mood: phosphorescent

[*] posted on 28-4-2016 at 00:06


Hi Dhaffnersr

The values I get are :-

Fluorescein - 528nm
Eosin Y - 551nm
Rhodamine B - 577nm

All dyes dissolved in basic (0.1M NaOH ethanol)




If you're not part of the solution, you're part of the precipitate.
View user's profile View All Posts By User
dhaffnersr
Harmless
*




Posts: 30
Registered: 29-3-2016
Location: Hopewell VA
Member Is Offline

Mood: always cool and calm...

[*] posted on 28-4-2016 at 03:34


Not too bad ezza, I see you are using the NaOH/ethanol buffer, I have been experimenting with that because of the higher pH. My solution is 0.1M NaOH and 20% ethanol with a pH of 12, I did have to adjust the sample concentrations though, for the Rhodamine B I had to use the 100ppm an 100ppb standard, for the Eosin Y- samples 1 thru 6 were-58microlitres to 372microlitres, for the fluorescein it was 11microlitres to 77microlitres.

My peaks were- fluorescein-517nm-543nm, Eosin Y-543nm, Rhodamine B-565nm at 100ppm.

here is a great reference site to use for comparisons-

http://omlc.org/spectra/PhotochemCAD/html/010.html

They have many spectra with raw data included.
View user's profile Visit user's homepage View All Posts By User