Catalytic Hydrogenation comments......
I've been trying to figure out how much hydrogen a reaction will take up when all one has at disposition is a pressure gauge that one can use to count
the amount of psi's used up in the reaction,.........but how does that help in figuring out the total amount of hydrogen used up.....I would like to
figure this one out as my Parr Hydrogenator is very useful but I want to calculate amount usage vs. catalyst poisoning and needing to replenish the
catalyst as opposed to thinking the reaction has been fully hydrogenated.........solo
It's better to die on your feet, than live on your knees....Emiliano Zapata.
|
|
smuv
National Hazard
Posts: 842
Registered: 2-5-2007
Member Is Offline
Mood: Jingoistic
|
|
Look at this. The only thing slightly inconvenient about this is you need to know with reasonable accuracy the volume of the headspace above the
hydrogenation solvent.
So it would be something like moles used = moles initial - moles final.
|
|
Magpie
lab constructor
Posts: 5939
Registered: 1-11-2003
Location: USA
Member Is Offline
Mood: Chemistry: the subtle science.
|
|
Solo,
PV=nRT should be useful here. I'm assuming you know what this means. If not let me (us) know.
The single most important condition for a successful synthesis is good mixing - Nicodem
|
|
solo
International Hazard
Posts: 3975
Registered: 9-12-2002
Location: Estados Unidos de La Republica Mexicana
Member Is Offline
Mood: ....getting old and drowning in a sea of knowledge
|
|
Actually I would like to hear from members that are familiar with my dilemma ......those that have done catalytic hydrogenations and have figured
exactly how much hydrogen will be taken up and have managed to know that by the only parameter available on the equipment....a pressure gauage! I
have in the past, while doing similar reduction with similar molar content, been able to estimate the number of psi that need to be taken up before
the reaction was complete.........but that was by observation no real calculated parameters or values of psi to guide me, only after doing tlc was I
able to determine if the reaction was complete....... but then that's probably the only way to do this , hydrogenate till no more is taken up do a
tlc and then decide if the uptake stopped because the catalyst got poisoned or because the reaction is complete.......solo
It's better to die on your feet, than live on your knees....Emiliano Zapata.
|
|
smuv
National Hazard
Posts: 842
Registered: 2-5-2007
Member Is Offline
Mood: Jingoistic
|
|
Based on the formulas we have given you, if you know the volume of gas in the pressurized vessel, its temp and the pressure you can calculate the
number of moles of gas in the vessel. So I guess you need to invest in a thermometer and a ruler to measure the temp and volume accordingly.
Any formula that does not rely on temperature makes no sense (assuming the temp is not fixed). Any formula that calculates moles without a volume
makes no sense. If you just think about it for a minute, you should understand that there is no logical way of calculating the moles of gas without
pressure,volume and temperature.
[Edited on 1-2-2008 by smuv]
|
|
S.C. Wack
bibliomaster
Posts: 2419
Registered: 7-5-2004
Location: Cornworld, Central USA
Member Is Offline
Mood: Enhanced
|
|
Perhaps they did things differently than you think they did? Perhaps they calculated the theoretical amount of hydrogen necessary, but used the gauge
only to tell when no more hydrogen was being absorbed? Perhaps while they were at it, they also calculated yields based on weight gain or isolated
product? Maybe you could give us an example, a quote?
|
|
Nicodem
Super Moderator
Posts: 4230
Registered: 28-12-2004
Member Is Offline
Mood: No Mood
|
|
Estimating the PSI/mmol value by using PV=nRT is one way of doing it.
The other one is doing a calibration with an easily reducible alkene (styrene, cyclohexene, etc.) using the same volume of solvent as in later
reductions (and preferably the same solvent anyway). I did this once on my Parr, just to get a rough picture of how much the PSI/mmol value is. I
don't see any reason to get an accurate measurement. After all, you generally don't use the same amount of solvent for every hydrogenation you do.
If I remember correctly, there are also some estimates on the Parr institute homepage (assuming you use their standard size vessels for shaker type
apparatuses).
…there is a human touch of the cultist “believer” in every theorist that he must struggle against as being
unworthy of the scientist. Some of the greatest men of science have publicly repudiated a theory which earlier they hotly defended. In this lies their
scientific temper, not in the scientific defense of the theory. - Weston La Barre (Ghost Dance, 1972)
Read the The ScienceMadness Guidelines!
|
|
chemrox
International Hazard
Posts: 2961
Registered: 18-1-2007
Location: UTM
Member Is Offline
Mood: LaGrangian
|
|
You're working with two pressures. The pressure being supplied to the reaction vessel and the pressure on the supply tank. The pressure on the tank
is supposed to account for the use given constant temperature. The smaller the supply tank the easier its going to be to callibrate the process.
Another approach to consider, I have not done this, is to weigh the supply tank before and after the reaction. Weigh the H2 tank before connecting
it. As per usual, watch for the pressure to stop changing, disconnect the tank and weigh it again. Temperature no longer a worry this way and it
offers another way of callibrating. All you need is a balance with precision in the range of the tank weight. Physicists and engineers tend to have
things like that.
My guages don't have the kind of precission one would need for what you want. Given a large face, finely marked guage, you could keep a thermometer
in line with the gas and account for any T change. This would take you back to the gas law method. It should work out the same.
"When you let the dumbasses vote you end up with populism followed by autocracy and getting back is a bitch." Plato (sort of)
|
|
solo
International Hazard
Posts: 3975
Registered: 9-12-2002
Location: Estados Unidos de La Republica Mexicana
Member Is Offline
Mood: ....getting old and drowning in a sea of knowledge
|
|
Well I've given up in trying to calculate the amount of hydrogen used based on the PSI's used, since it depends on head space , solvent......so I
continue to use the method I've used for many years.....hydrogenate till no more.
Recently I've had to hydrogenate some experimental material that doesn't take hydrogen so rapid and in the literature the hydrogenation can take 18
hours at nl pressure, I would suppose that under a higher pressure the reaction should take hydrogen much faster ....... but it doesn't, another thing
I've noticed by using the wet Pd/C the actual amount of Pd is reduced to 50% by weight, true it saves the experimenter from flame ups and can be
shipped by plane since there is no risk but it doesn't seem to work as well as the dry Pd/C that I've used for so many years. I also found that the
wet Pd seems to get poisoned much easier.....
Getting back to why some reactions that claim to be catalytic hydrogenation reactions take so long.....inpatient........solo
It's better to die on your feet, than live on your knees....Emiliano Zapata.
|
|
Drunkguy
Hazard to Others
Posts: 172
Registered: 23-12-2005
Member Is Offline
Mood: somewhat pissed.
|
|
You have ur own Parr Hydrogenator? That's not bad going.
I bought a Manometer for similar reasons but dont think thats much good.
I think I read that the reaction is completed when the pressure doesnt keep dropping every few hours and stays constant.
I thought the way you can work out if the theoretical amount of hydrogen has been absorbed must be done using a volumetric technique. For instance a
burette is partially filled with water and the reading before the reaction is noted. After the reaction the reading will have changed.
Infact, in Vogel it shows 2 burettes hooked up so that it works as a manometer. That is for atmospheric pressure hydrogenation though where u
calculate it based on volume.
Maybe if you know the volume of your vessel, then you can plug the pressure drop recording into an equation and use the data to calculate amount of
hydrogen absorbed?
Man, you gott read some of those books like Rylander etc.
|
|
WizardX
Hazard to Self
Posts: 61
Registered: 11-8-2005
Location: wizardx.4shared.com
Member Is Offline
Mood: wizardx.suddenlaunch3.com
|
|
Standardization: http://www.parrinst.com/doc_library/members/271m.pdf
The apparatus can be standardized by making a preliminary run with a known amount of any compound that can be completely and quantitatively reduced.
The pressure drop per mole of hydrogen
consumed in such tests is then used as a basis for estimating the progress of a reaction with unfamiliar materials. A procedure is described in
reference (3) for standardizing the apparatus by reducing
11.6 grams (0. 1 mole) of pure maleic acid dissolved in 150 mL of 95% ethanol using 0.1 gram of catalyst. The reaction is carried out as previously
described with shaking continued until no more hydrogen
is consumed. This usually takes twenty to thirty minutes, after which the pressure drop in the tank is recorded. Since exactly 0.1 mole of hydrogen
has been consumed in this run, this decrease in
tank pressure can be used as a basis for measuring or regulating the amount of hydrogen consumed when treating other compounds.
APPENDIX F. Molar amount of each gas per batch in pdf file: BATCH AQUEOUS-PHASE REFORMING OF LIGNOCELLULOSIC. BIOMASS FOR HYDROGEN PRODUCTION
http://etd.gatech.edu/theses/available/etd-07102006-141653/u...
Albert Einstein - \"Great ideas often receive violent opposition from mediocre minds.\"
|
|
Drunkguy
Hazard to Others
Posts: 172
Registered: 23-12-2005
Member Is Offline
Mood: somewhat pissed.
|
|
^Yeah, that sounds like a very sensible way of doing it.
It's important that you have an accurate analytical piece of apparatus to record the pressure drop though. If it's just a gauge then im not sure if it
would allow you to do it with that level of precision.
|
|
Nicodem
Super Moderator
Posts: 4230
Registered: 28-12-2004
Member Is Offline
Mood: No Mood
|
|
If you are doing hydrogenations that consume more than just a couple mmols of hydrogen then the normal gauge showing deltaP of 1 PSI is just fine.
Like I already wrote somewhere above, I did such a calibration once using acetophenone (takes up 2 mol H2 per mol substrate, getting reduced to
ethylbenzene). If I remember correctly it took approximately 2 PSI per mmol hydrogen for a 100ml filled 250ml Parr bottle. Such calibrations are
important if you do hydrogenations selective on one double bond over some other, otherwise you just hydrogenate till constant pressure. However,
sometimes summing up the pressure drop helps to estimate if a reduction is not going as it should or that the catalyst got poisoned.
|
|
solo
International Hazard
Posts: 3975
Registered: 9-12-2002
Location: Estados Unidos de La Republica Mexicana
Member Is Offline
Mood: ....getting old and drowning in a sea of knowledge
|
|
Within the scope of problems with catalytic hydrogenations....there is a recurrent problem with poisoning of the Palladium catalyst....an older
chemist suggested to a friend that it lies in the imine, in the case where reductive amination is being employed...., it was suggested to add some
activated charcoal to the imine solution and filter then add the desired Pd catalyst and hydrogenate, without the fear of the catalyst being
poisoned.......and perhaps it may be the reason why only 50 % of the imine in reductive amination of phenylacetone stops there and won't go any
further....or in that case is there a different reason ....?.........solo
It's better to die on your feet, than live on your knees....Emiliano Zapata.
|
|
Barium
Hazard to Self
Posts: 85
Registered: 24-8-2008
Member Is Offline
Mood: No Mood
|
|
Palladium should not be the first choice of catalyst for that reduction to begin with. Platinum is much more effective. Different carriers can change
the activity and selectivity as well. Carbon is the normal choice but acidic alumina usually gives great yields in this type of hydrogenations.
Activated carbon should suffie to remove any poisons before the hydrogenation commences. If sulfur is the suspected poison, then a little raney nickel
is the best choice since it will form NiS irreversibly. After removal of the nickel the noble metal catalyst can be added and the process started as
usual.
Save any old inactive raney nickel for this very purpose instead of destroying it.
|
|
solo
International Hazard
Posts: 3975
Registered: 9-12-2002
Location: Estados Unidos de La Republica Mexicana
Member Is Offline
Mood: ....getting old and drowning in a sea of knowledge
|
|
Quote: | For example, desoxyephedrine was synthesized by reductive alkylation of phenyl-2-propanone and methylamine in the presence of hydrogen and palladium
on carbon. Mixing of an equal volume (mole ratio 1:1.7) of P2P (7) and 40% methylamine in methanol immediately produced the imine (8) shown in Fig. 2.
After 24 h, desoxyephedrine (9) is the main component (50%). The mass spectrum of the imine compound, 1-phenyl-2-methyl-2-(methylimino)-propane, is
shown in Fig. 3. The imine compound is fairly stable since it does not readily reduce under the reaction conditions (50 lb hydrogen, Pd/C).
|
........................source,
http://psychedelichosting.info/Ionium/Rhodium/chemistry/meth...
.........so then the remaining imine that can't be reduced via Pd/C hydrogenation lends itself to be processed by using PtO Adam's Catalyst?....or is
there a way to improve the reaction using Raney Nickle ? .....nice to see you here Barium , this is Java from the
Hive.................solo
[Edited on 30-8-2008 by solo]
It's better to die on your feet, than live on your knees....Emiliano Zapata.
|
|
Barium
Hazard to Self
Posts: 85
Registered: 24-8-2008
Member Is Offline
Mood: No Mood
|
|
Thank you Solo/Java
In that article the author did not specify which catalyst he used. So it is really hard to comment on that. But I can guarantee that way better yield
and shorter reaction times would be the result by using, pretty much, any type of Platinum on carbon. For God's sake don't use Adams catalyst, unless
you have enough money to whipe you ass with. Always use prescious metal catalysts deposited on a carrier. I know university labs still use those old
catalysts but that is either because they (a) don't know any better, or (b) have too much money on their hands.
If you can make a really active Raney nickel you should be able to use it at such low pressures for the reduction of a imine.
|
|
Vogelzang
Banned
Posts: 662
Registered: 26-4-2008
Member Is Offline
Mood: No Mood
|
|
nickel catalysts.
https://www.hyperlab.info/inv/index.php?act=ST&f=17&...
|
|
Barium
Hazard to Self
Posts: 85
Registered: 24-8-2008
Member Is Offline
Mood: No Mood
|
|
I have never had any success with any Urushibara type of catalyst. I would sure like to know why since I tried several different types of catalyst on
several different functional groups. The insensitivity of the catalysts towards storage and ease of preparation made them highly interesting. But the
lack of success made me somewhat puzzled.
|
|
Epopteia
Harmless
Posts: 8
Registered: 19-7-2005
Member Is Offline
Mood: Yes
|
|
Yep, I've also tried many different methods for using Urushibara catalysts, none of which worked well at all. The difficulties in getting them to work
led me to pursue other types of reducing systems.
|
|
gsd
National Hazard
Posts: 847
Registered: 18-8-2005
Member Is Offline
Mood: No Mood
|
|
I have got excellent results with Urushibara - A catalyst.
For reduction of C = O to C- H.lI have used 3 different catalysts systems:
1) 10 % Pd/C : which was best for the job. I could recycle the catalyst 10 times without any problem. The temperature and pressure conditions were
also very mild.
2) U-Ni-A : Was nearly as good, could be recycled 10 times but T/P conditons were higher as compared to Pd/C
3) Raney Ni: Could complete the reduction at all. Inspite of using drastic T/P conditions , the reduction proceeded only upto C - OH
gsd
|
|
Klute
International Hazard
Posts: 1378
Registered: 18-10-2006
Location: France
Member Is Offline
Mood: No Mood
|
|
Was this the desoxygenation of aromatic ketones, or any aliphatic ketone?
What kind of pressures and temps were used?
\"You can battle with a demon, you can embrace a demon; what the hell can you do with a fucking spiritual computer?\"
-Alice Parr
|
|
Nicodem
Super Moderator
Posts: 4230
Registered: 28-12-2004
Member Is Offline
Mood: No Mood
|
|
Only aromatic ketones (Ar-CO-R; where R is alkyl not containing any heteroatoms at the alpha position) can be hydrogenated to the methylene group
(A-CH2-R). This is generally done with Pd-C at very mild conditions (even 1 atm H2 pressure at rt works). For, example acetophenone is reduced in less
than one minute in methanol with ~1 mol% 5% Pd-C at ~2 bar of H2 pressure. Up to now I was not aware Urushibara nickel can do the same (Gsd, can you
describe the conditions you used in more detail?), but I do know Raney nickel can't (it is actually used to reduce ketones to alcohols).
Aliphatic ketones can't be reduced that way. For their reduction via hydrogenation you can make the thioacetals (using HSCH2CH2SH or similar) and
desulfurize with Raney nickel, but that is not particularly practical.
…there is a human touch of the cultist “believer” in every theorist that he must struggle against as being
unworthy of the scientist. Some of the greatest men of science have publicly repudiated a theory which earlier they hotly defended. In this lies their
scientific temper, not in the scientific defense of the theory. - Weston La Barre (Ghost Dance, 1972)
Read the The ScienceMadness Guidelines!
|
|
Klute
International Hazard
Posts: 1378
Registered: 18-10-2006
Location: France
Member Is Offline
Mood: No Mood
|
|
Thank you for the clarification, Nicodem. Now that i think of it, I think it can be done by CTH with Pd/C and formate too.
I guess for the complete desoxygenation of aliphatic ketones, a Wolf Kishner or Clemmensen is still in order...
\"You can battle with a demon, you can embrace a demon; what the hell can you do with a fucking spiritual computer?\"
-Alice Parr
|
|
stoichiometric_steve
National Hazard
Posts: 827
Registered: 14-12-2005
Member Is Offline
Mood: satyric
|
|
Klute, it can. There is a paper on it, although i dont have access to it right now. They used Pd/C with NH4COOH in hot GAA. The yields weren't too
good, though.
|
|