Sciencemadness Discussion Board

Data Validation of previous Eosin Y results for April/21/2016

dhaffnersr - 23-4-2016 at 11:18

Using my Aries series green laser, I wanted to validate the Eosin Y study I did the other day with the Prolight Uv LED (410nm). I'm presenting this study as a final result for this series of tests using my plab v2.5 spectrometer (build 3.0)

Sample preparation:
2 cuvettes, one for the Blank (0.1M NaOH in 20 % ethanol) and one for samples S1 thru S6.
Volume capacity calculations- 1 drop of eosin y (1 % aqueous solution) = 0.058g
To convert to micro liters take 0.058g*1000 = 58 microliters
Sample(1)-1 drop 58 microliters
Sample(2)-2 drops 127 microliters
Sample(3)- 3 drops 189 microliters
Sample(4)- 4 drops 254 microliters
Sample(5)- 5 drops 320 microliters
Sample(6)- 6 drops 372 microliters
Each scan will last 30 seconds.
Excitation source- 532nm Aries series green laser (20mW)



green laser3.png - 829kB PLOT 1 eosin Y apr22.png - 187kB PLOT 2 eosin Y apr22.png - 256kB emmssn eosin y.png - 27kB 2rawdata eosiny S1 apr22.png - 41kB S1 thru S6 532nm eosin y apr22.png - 34kB S1 thru S6 with BLANK plot apr22.png - 32kB

dhaffnersr - 23-4-2016 at 13:52

More on my research can be found here:

http://publiclab.org/profile/dhaffnersr

;)


nezza - 24-4-2016 at 23:47

Interesting work, and some of your publiclab studies look good as well. I have recently been doing some spectrometry (Not as comprehensively as yourself) on some fluorescent dyes including Eosin Y. The spectrometer I use is a science-surplus unit using spectrum studio software. Using a white light source to generate an absorption spectrum I find a spectrum shift on fluorescent materials with concentration. I am assuming this is due to some self fluorescence from the dye caused by the white light source. I know Eosin's maximum absorption is in the green so I will try to obtain an emission spectrum of the dye. I have attached a plot of the curve at different concentrations. I suspect a scanning spectrophotometer using monochromatic light would be the best way to obtain a true absorption spectrum.

Eosin.jpg - 144kB

dhaffnersr - 25-4-2016 at 02:34

Hey nezza, nice plot. Remember that at higher sample concentrations your going to start suffering the "inner-filter" effect, I usually start with 6 different sample concentrations, this usually gives the best calibration curves.

Yes, you are correct about using the monochromatic light, I use a UV laser (405nm) and a green laser (532nm) for my excitation light sources.:cool:

nezza - 26-4-2016 at 23:21

Thanks for the info. I have attached my traces of 3 dyes using a 405nm laser for the fluorescein and a 532nm laser for the other 2. Th traces are much cleaner.

Dyes.jpg - 169kB

dhaffnersr - 27-4-2016 at 03:36

hey nezza:cool:

very clean, what are the peak wavelengths? curious because I have to use Plab's spectral workbench 2.0 software online to run my hardware to capture spectra and they run java script, and I think there are inconsistencies in my results due to the way they average the pixels.

nezza - 28-4-2016 at 00:06

Hi Dhaffnersr

The values I get are :-

Fluorescein - 528nm
Eosin Y - 551nm
Rhodamine B - 577nm

All dyes dissolved in basic (0.1M NaOH ethanol)

dhaffnersr - 28-4-2016 at 03:34

Not too bad ezza, I see you are using the NaOH/ethanol buffer, I have been experimenting with that because of the higher pH. My solution is 0.1M NaOH and 20% ethanol with a pH of 12, I did have to adjust the sample concentrations though, for the Rhodamine B I had to use the 100ppm an 100ppb standard, for the Eosin Y- samples 1 thru 6 were-58microlitres to 372microlitres, for the fluorescein it was 11microlitres to 77microlitres.

My peaks were- fluorescein-517nm-543nm, Eosin Y-543nm, Rhodamine B-565nm at 100ppm.

here is a great reference site to use for comparisons-

http://omlc.org/spectra/PhotochemCAD/html/010.html

They have many spectra with raw data included.